Filters
Vendor
Field Strength

Seizures Brain 3T

Philips Healthcare – MR Gadopiclenol (VUEWAY®) Protocols

Get Adobe Reader

Instructions And Patient Preparation

Instructions
• Change patient into hospital issued gown and pants
• 20- to 22-gauge IV access preferred
Position
Supine, head first
Coil
32 channel head coil
Landmark
Glabella
Comments
The Coronal T2 FLAIR and T2 sequences should be acquired perpendicular to the plane of the hippocampus

Contrast Agent Parameters

Contrast Agent Dose Injection Rate
Gadopiclenol (VUEWAY) 0.05 mmol/kg Hand inject
Saline Flush 10-20 mL Hand inject
Comments The Coronal T2 FLAIR and T2 sequences should be acquired perpendicular to the plane of the hippocampus

Info Page

Scan Description Survey
Brain
SAGITTAL
3D T1
SAGITTAL
3D T2 FLAIR
CORONAL
T2 Hipp
CORONAL
T2 FLAIR Hipp
AXIAL
Gray Matter
SWIp
Fast Bold
Source
DTI
SAGITTAL
3D T1 +C
AXIAL
T1 +C Fat Sat
  View Image View Image View Image View Image     View Image View Image View Image View Image
Total scan duration 00:31.6 02:21.7 05:20.0 01:48.0 03:40.0 05:44.6 03:05.2 02:00.8 02:21.7 03:21.7
Rel. SNR 1.74 1.28 0.61 0.71 0.59 1.55 0.83 1.35 1.28 0.81
Act. TR (ms) — — — 3000 — — — 4923 — 580
Act. TR/TE (ms) 15 / 5.2 7.7 / 3.5 — — — — — — 7.7 / 3.5 —
Act. TR/TE1/delta TE (ms) — — — — — — 31 / 7.2 / 6.2 — — —
Act. TR/TI (ms) — — 8000 / 2400 — 11000 / 2800 15662 / 3400 — — — —
Act. TE (ms) — — 363 80 120 25 — 68 — 10
ACQ matrix M x P 256 x 126 220 x 220 240 x 240 312 x 236 256 x 162 240 x 142 384 x 314 112 x 112 220 x 220 276 x 180
ACQ voxel MPS (mm) 0.98 / 1.98 / 10.0 1.09 / 1.09 / 1.00 1.00 / 1.00 / 1.20 0.58 / 0.76 / 3.00 0.70 / 1.11 / 3.00 0.96 / 1.29 / 3.00 0.60 / 0.60 / 2.00 2.00 / 2.00 / 3.00 1.09 / 1.09 / 1.00 0.80 / 0.99 / 3.00
REC voxel MPS (mm) 0.98 / 0.98 / 10.0 0.94 / 0.94 / 1.00 0.50 / 0.50 / 0.60 0.42 / 0.42 / 3.00 0.35 / 0.35 / 3.00 0.90 / 0.90 / 3.00 0.45 / 0.45 / 1.00 1.00 / 1.00 / 3.00 0.94 / 0.94 / 1.00 0.43 / 0.43 / 3.00
Scan percentage (%) 49 100 100 76 63 74 100 100 100 80
Packages — — — 4 4 2 — 1 — 12
Min. slice gap (mm) — — — 3 3 3 — 0 — 3
Act. slice gap (mm) — 0 -0.6 — — — -1 — 0 —
Optimal slices — — — 14 24 — — — — 8
Max. slices — — — 28 48 — — — — —
TFE shots 3 70 — — — — — — 70 —
TFE dur. shot / acq (ms) 701.6 / 630.0 1740.9 / 1695.7 — — — — — — 1740.9 / 1695.7 —
TFE shot interval (ms) 701.6 — — — — — — — — —
Min. TI delay 375.1 884.6 — — — — — — 884.6 —
Act. WFS (pix) / BW (Hz) 2.322 / 187.1 2.080 / 208.9 — — — — 1.702 / 255.3 — 2.080 / 208.9 —
Min. WFS (pix) / Max. BW (Hz) 1.050 / 413.8 0.482 / 901.9 — — — — 0.761 / 571.1 — 0.482 / 901.9 —
Min. TR/TE (ms) 10 / 5.2 — — — — — — — — —
Min. TR/TE1/delta TE (ms) — — — — — — 29 / 5.8 / 6.2 — — —
EPI factor — — — — — — — 45 — —
WFS (pix) / BW (Hz) — — 0.801 / 542.5 2.136 / 203.4 1.993 / 218.0 1.714 / 253.4 — 13.193 / 32.9 — 2.005 / 216.7
BW in EPI freq. dir. (Hz) — — — — — — — 1844.5 — —
Full flow comp. — — no — — — — — — yes
TSE es / shot (ms) — — 5.9 / 685 10.0 / 160 10.9 / 229 6.9 / 117 — — — 10.0 / 30
TEeff / TEequiv (ms) — — 363 / 166 80 / 72 120 / 106 — — — — —
Min. TR (ms) — — — 2419 — — — — — 580
Min. TR/TI (ms) — — 3098 / 50 — 10218 / 50 — — — — —
SPAIR TR (ms) — — 8000 — — — — — — —
SPAIR inv. delay (ms) — — 220 — — — — — — —
SPAIR offset act./default (Hz) — — 100 [220] — — — — — — —
Local torso SAR < 31 % < 20 % < 8 % < 81 % < 65 % < 75 % < 8 % < 83 % < 20 % < 100 %
Whole body SAR / level < 0.7 W/kg / normal < 0.5 W/kg / normal < 0.2 W/kg / normal < 1.9 W/kg / normal < 1.5 W/kg / normal < 2.4 W/kg / 1st level < 0.3 W/kg / normal < 1.9 W/kg / normal < 0.5 W/kg / normal < 2.3 W/kg / 1st level
SED 0.0 kJ/kg < 0.1 kJ/kg < 0.1 kJ/kg < 0.2 kJ/kg < 0.3 kJ/kg < 0.8 kJ/kg 0.0 kJ/kg < 0.2 kJ/kg < 0.1 kJ/kg < 0.5 kJ/kg
Coil Power — 10% 4% 38% 31% 72% 8% 39% 10% 48%
Max B1+rms 0.90 uT 0.73 uT 0.49 uT 1.47 uT 1.35 uT 2.12 uT 0.65 uT 1.47 uT 0.73 uT 1.62 uT
PNS / level 17 % / normal 55 % / normal 60 % / normal 20 % / normal 24 % / normal 49 % / normal 79 % / normal 86 % / 1st level 55 % / normal 57 % / normal
dB/dt 15.4 T/s 53.1 T/s 86.7 T/s 15.4 T/s 15.4 T/s 46.4 T/s 55.3 T/s 112.7 T/s 53.1 T/s 87.1 T/s
Sound Pressure Level (dB) 3 20 22 4 6 18 24 22 20 18

Geometry

Scan Description Survey
Brain
SAGITTAL
3D T1
SAGITTAL
3D T2 FLAIR
CORONAL
T2 Hipp
CORONAL
T2 FLAIR Hipp
AXIAL
Gray Matter
SWIp
Fast Bold
Source
DTI
SAGITTAL
3D T1 +C
AXIAL
T1 +C Fat Sat
                     
SmartSelect yes yes yes yes yes yes yes yes yes yes
Coil 1 (exclude) — None None None None None None None None None
Uniformity — CLEAR CLEAR CLEAR CLEAR CLEAR CLEAR CLEAR CLEAR CLEAR
FOV AP (mm) 250 240 240 95 175 230 230 224 240 220
      FOV (mm) (3) 50, (17) 10 — — — — — — — — —
FOV FH (mm) 250 240 240 180 180 150 150 150 240 144
FOV RL (mm) — 185 156 180 180 183 189 224 185 179
ACQ voxel size AP (mm) 2 1 1 — — 1 0.6 2 1 0.8
ACQ voxel size FH (mm) 1 1 1 0.6 0.7 — 2 — 1 —
ACQ voxel size RL (mm) — 1 1.2 0.7 0.9 1.3 0.6 2 1 1
Slice thickness (mm) 10 — — 3 3 3 — 3 — 3
Bar orientation RL — — — — — — — — —
VOI orientation transverse — — — — — — — — —
VOI size AP (mm) 30 — — — — — — — — —
VOI size FH (mm) 30 — — — — — — — — —
VOI size RL (mm) 30 — — — — — — — — —
Recon voxel size AP (mm) 1 0.9 0.5 — — 0.9 0.4 1 0.9 0.4
Recon voxel size FH (mm) 1 0.9 0.5 0.4 0.4 — 1 — 0.9 —
Recon voxel size RL (mm) — 1 0.6 0.4 0.4 0.9 0.4 1 1 0.4
Fold-over suppression no no no no oversampling no no no no no
       A (mm) 50 — — — — — — — — —
       L (mm) — — — — 10 — — — — —
       P (mm) 50 — — — — — — — — —
       R (mm) — — — — 10 — — — — —
Slice oversampling — default user defined — — — default — default —
       oversample factor — — 1 — — — — — — —
ENCASE enable — no — — — — no — no —
Reconstruction matrix 256 256 480 432 512 256 512 224 256 512
SENSE no no no no no no no yes no no
       P reduction (AP) — — — — — — — 2.5 — —
MB SENSE — — — — — — — no — —
CS-SENSE no yes yes yes yes no yes no yes yes
       reduction — 3 6 2 2.3 — 4.5 — 3 2.3
       denoising — system default system default system default system default — system default — system default system default
k-t Acceleration no no no no no no no no no no
Stacks 3 1 1 1 1 1 1 1 1 1
       type (20) parallel — — parallel parallel parallel — parallel — parallel
       slices (3) 3, (17) 1 185 260 24 44 50 150 50 185 48
       slice gap user defined — — user defined user defined user defined — user defined — user defined
       gap (mm) (20) 10 — — 1 1 0 — 0 — 0
       slice orientation sagittal, coronal, (18) transverse sagittal sagittal coronal coronal transverse transverse transverse sagittal transverse
       fold-over direction AP, (19) RL AP AP RL RL RL RL AP AP RL
       fat shift direction (2) F, (18) P F F F F P P P F P
       radial axis AP, RL, AP, (17) RL — — — — — — — — —
       radial angle (deg) (20) 0 — — — — — — — — —
Multi-chunk — no no — — — no — no —
Minimum number of packages — — — 1 3 1 — 1 — 1
Slice scan order default — — default default default — default — default
Stack scan order ascend — — — — — — — — —
Move table per stack no — — — — — — — — —
O-MAR — — no no no no — — — no
Large table movement no no no no no no no no no no
Stack alignment no — — — — — — — — —
Stack sort order no — — — — — — — — —
PlanAlign no no no no no no no no no no
REST slabs 0 0 0 0 0 0 0 0 0 0
       orientation — — — — — — — transverse — —
Patient position head first head first head first head first head first head first head first head first head first head first
Patient body position head first head first head first head first head first head first head first head first head first head first
Patient orientation supine supine supine supine supine supine supine supine supine supine
Patient body orientation supine supine supine supine supine supine supine supine supine supine

Contrast

Scan Description SurveyBrain SAGITTAL3D T1 SAGITTAL3D T2 FLAIR CORONALT2 Hipp CORONALT2 FLAIR Hipp AXIALGray Matter SWIpFast Bold SourceDTI SAGITTAL3D T1 +C AXIALT1 +C Fat Sat
Scan type Imaging Imaging Imaging Imaging Imaging Imaging Imaging Imaging Imaging Imaging
Scan mode M2D 3D 3D MS MS MS 3D MS 3D MS
3D non-selective — no — — — — no — no —
Contrast enhancement T1 T1 — — — — T1 — T1 —
Modified SE — — — no — — — no — no
Acquisition mode cartesian cartesian cartesian cartesian cartesian cartesian cartesian cartesian cartesian cartesian
Fast Imaging mode TFE TFE TSE TSE TSE TSE none EPI TFE TSE
3D VIEW — — no — — — — — — —
TSE factor — — 110 15 21 17 — — — 3
TFE factor 42 220 — — — — — — 220 —
Echoes 1 1 1 1 1 1 4 1 1 1
TE shortest shortest shortest user defined user defined user defined user defined shortest shortest user defined
Flip angle (deg) 20 8 — 90 — — 17 90 8 90
Refocusing control — — constant constant constant no — — — no
TR user defined shortest user defined user defined user defined shortest user defined shortest shortest range
Halfscan no no — no no yes no yes no no
Water-fat shift maximum maximum maximum maximum user defined maximum user defined minimum maximum maximum
IR delay (ms) — — 2400 — 2800 3400 — — — —
RF Shims — fixed fixed fixed fixed fixed fixed fixed fixed fixed
Shim — auto default default default auto auto PB-volume auto default
ShimAlign — — — — — — — no — —
mDIXON no no no no no no no no no no
Fat suppression no no SPAIR no SPIR SPIR no SPIR no SPIR
Grad Rev Fat suppr — — — — — — — no — —
Water suppression no no no no no no no no no no
BB pulse — — no no no no — no — no
TFE prepulse invert invert — — — — — — invert —
MTC no no no no no no no no no no
Diffusion mode no no no no no no no DTI no no
T1 mapping no no no no no no no no no no
SAR mode high high high high high high high high high high
B1 mode default default default default default default default default default default
SAR allow first level yes yes yes yes yes yes yes yes yes yes
PNS mode low moderate high low low low moderate high moderate high
Gradient mode regular default maximum default default default maximum maximum default maximum
SofTone mode no no no user defined user defined no no no no no

Motion

Scan Description Survey
Brain
SAGITTAL
3D T1
SAGITTAL
3D T2 FLAIR
CORONAL
T2 Hipp
CORONAL
T2 FLAIR Hipp
AXIAL
Gray Matter
SWIp
Fast Bold
Source
DTI
SAGITTAL
3D T1 +C
AXIAL
T1 +C Fat Sat
                     
Cardiac synchronization no no no no no no no no no no
Respiratory compensation no no no no no no no no no no
Navigator respiratory comp no no no no no no no no no no
Flow compensation no no sensitized no no no yes no no yes
      direction — — — — — — — — — in-plane
Temporal slice spacing — — — default — — — default — default
fMRI echo stabilisation no no — — — — no — no —
Motion smoothing — — no yes no no — — — no
NSA 1 1 1 1 1 2 1 1 1 1
SMART — — — — — no — — — —
MRE enable no no no no no no no no no no

Dynamic Angio

Scan Description Survey
Brain
SAGITTAL
3D T1
SAGITTAL
3D T2 FLAIR
CORONAL
T2 Hipp
CORONAL
T2 FLAIR Hipp
AXIAL
Gray Matter
SWIp
Fast Bold
Source
DTI
SAGITTAL
3D T1 +C
AXIAL
T1 +C Fat Sat
                     
Angio / Contrast enh. no no — — — — no — no —
Quantitative flow no no — — — — no — no —
CENTRA — no no — — — — — no —
Manual start no no no no no no no no no no
Dynamic study no no no no no no no no no no
      dyn stabilization — — — — — — — regular — —
Arterial Spin labeling no no no no no no no no no no

Post Processing

Scan Description Survey
Brain
SAGITTAL
3D T1
SAGITTAL
3D T2 FLAIR
CORONAL
T2 Hipp
CORONAL
T2 FLAIR Hipp
AXIAL
Gray Matter
SWIp
Fast Bold
Source
DTI
SAGITTAL
3D T1 +C
AXIAL
T1 +C Fat Sat
                     
Preparation phases — auto auto auto auto auto auto auto auto auto
Interactive F0 no no no no no no no no no no
Quick Survey default default default default default default default default default default
B0 field map no no no no no no no no no no
MIP/MPR no no no no no no no no no no
SWIp no no — — — — yes — no —
Images M, (3) no M, (3) no M, (3) no M, (3) no M, (3) no M, (3) no SW-M, SW-P, (2) no M, (3) no M, (3) no M, (3) no
Autoview image M M M M M M — M M M
Calculated images (4) no (4) no — (4) no — — — (4) no (4) no (4) no
Reference tissue White matter White matter Grey matter Grey matter White matter Grey matter White matter Grey matter White matter Grey matter
Recon compression No No No No No No High No No No
Preset window contrast soft soft soft soft soft soft soft soft soft soft
Reconstruction mode immediate immediate real time real time real time real time immediate immediate immediate real time
Save raw data no no no no no no no no no no
Hardcopy protocol no no no no no no no no no no
Image filter system default system default system default system default system default system default system default system default system default system default
Uniformity correction no no no no no no no no no no
Geometry correction default default default default default default default default default default
Elliptical k-space shutter — default default — — — default — default —
EPI geometry correction — — — — — — — no — —
ADC correction — — — — — — — no — —
Free rotatable no no no no no no no no no no
IF_info_seperator 0 1634755923 1634755923 0 0 1634755923 1634755923 0 1634755923 1634755923