Filters
Vendor
Field Strength

MRA Neck 1.5T

Siemens Healthineers – MR Gadopiclenol (VUEWAY®) Protocols

Get Adobe Reader

Instructions And Patient Preparation

Instructions
These sequences are in addition to the Routine Adult Brain protocol
Position
Supine, head first
Coil
20 channel Head/Neck coil or better
Landmark
Glabella

Contrast Agent Parameters

Contrast Agent Dose Injection Rate
Gadopiclenol (VUEWAY) 0.05 mmol/kg TWIST CORONAL – Inject 1-2 mL contrast
Inject remaining contrast – Post CE MRA – 2 measurements
Saline Flush 20 mL 1-1.5 mL/sec
Comments • Report agent administered and dose
• Use minimal dose for patients with a GFR <60 mL/sec

Imaging Parameters

Scan Description MRA Neck
3D TOF
AXIAL
T1 Fat Sat
Black Blood
(When Requested)
Pre
CE MRA
CORONAL
TWIST
Inject 1-2 mL Contrast
Post
CE MRA
Inject Remaining Contrast
2 Measurements
  View Image View Image View Image View Image View Image
Sequence Name fl_r se fl fldyn fl
Dimension 3D Off 3D 3D 3D
Acceleration Mode GRAPPA ― GRAPPA GRAPPA GRAPPA
Acceleration Factor PE 2 ― 3 2 3
Acceleration Factor 3D 1 ― 1 1 1

Acquisition Parameters

Scan Description MRA Neck
3D TOF
AXIAL
T1 Fat Sat
Black Blood
Pre
CE MRA
CORONAL
TWIST
Post
CE MRA
           
TR 26.0 ms 627.0 ms 3.28 ms 2.76 ms 3.28 ms
TE 4.87 ms 17.00 ms 1.26 ms 1.03 ms 1.26 ms
Flip Angle 20 deg 90 deg 25 deg 25 deg 25 deg
Bandwidth 103 Hz/Px 150 Hz/Px 590 Hz/Px 650 Hz/Px 590 Hz/Px
Echo Spacing 13.48 ms ― ― ― ―
Fat-Water Contrast Standard Fat Saturation Standard Standard Standard
Fat Saturation ― Strong ― ― ―
Gradient Mode Fast Fast Fast Fast Fast
RF Pulse Type Normal Normal Normal Normal Normal
Asymmetric Echo Weak Allowed Allowed Strong Allowed

Scan Parameters

Scan Description MRA Neck
3D TOF
AXIAL
T1 Fat Sat
Black Blood
Pre
CE MRA
CORONAL
TWIST
Post
CE MRA
           
Base Resolution 320 320 384 256 384
Phase Resolution 65% 80% 82% 85% 82%
Slice Resolution 63% ― 61% 54% 61%
Averages 1 1 1 ― 1
Concatenations 8 2 ― ― ―
FoV Read 190 mm 230 mm 370 mm 400 mm 370 mm
FoV Phase 75% 50% 81% 75% 81%
Phase Encoding Dir. A >> P A >> P R >> L R >> L R >> L
Slice Oversampling 23% ― 13% 20% 13%
Phase Oversampling 30% 80% 0% 0% 0%
Slice Partial Fourier 6/8 ― 6/8 6/8 6/8
Phase Partial Fourier Off ― Off 6/8 Off
Flow Compensation Slice/Read ― ― ― ―

Scanning Range

Scan Description MRA Neck
3D TOF
AXIAL
T1 Fat Sat
Black Blood
Pre
CE MRA
CORONAL
TWIST
Post
CE MRA
           
Elliptical Filter On On Off On Off
Normalize Prescan Prescan Prescan Prescan Prescan
Distortion Correction 2D 2D 2D 2D 2D
Distance Factor -35% 30% 20% ― 20%
Slice Thickness 1.0 mm 5.0 mm 0.8 mm 1.6 mm 0.8 mm
Slices 26 Slices per Slab 20 Slices 128 Slices per Slab 80 Slices per Slab 128 Slices per Slab
Time to center ― ― 6.0 s 8.8 s 6.0 s
Acquisition Time 6:18 min 4:56 min 24 sec 2:39 min 24 sec