Filters
Vendor
Field Strength

MR Venography Head 1.5T

Siemens Healthineers – MR Gadopiclenol (VUEWAY®) Protocols

Get Adobe Reader

Instructions And Patient Preparation

Instructions
• The following sequences are in addition to the Routine Adult Brain protocol
• MRA may be added for hyperacute stroke
Position
Supine, head first
Coil
20 channel Head/Neck coil or better
Landmark
Glabella

Contrast Agent Parameters

Contrast Agent Dose Injection Rate
Gadopiclenol (VUEWAY) 0.05 mmol/kg TWIST Coronal – Inject 1-2 mL contrast
Inject remaining contrast – Post CE MRV – 2 measurements
Saline Flush 20 mL 1-1.5 mL/sec (CEMRA) / 3-4mL/sec (DSC Perfusion)
Comments • For bifurcations. Noncontrast. 3D TOF if unable to do Carotid CE MRA or CE MRA is inadequate
• Noncontrast. VENC=15. Alternative to use Coronal plane with longer acquisition but more optimal raw data imaging
• Report agent administered and dose
• Use minimal dose for patients with a GFR <60 mL/sec

Imaging Parameters

Scan Description SAGITTAL
3D PC Venogram
Pre
CE MRV
CORONAL
TWIST
Inject 1-2 mL Contrast
Post
CE MRV
Inject Remaining Contrast
2 Measurements
  View Image View Image View Image View Image
Sequence Name fi fl fldyn fl
Dimension 3D 3D 3D 3D
Acceleration Mode GRAPPA GRAPPA GRAPPA GRAPPA
Acceleration Factor PE 2 2 2 2
Acceleration Factor 3D 1 2 2 2
AutoAlign ― Head > Basis Head > Basis Head > Basis

Acquisition Parameters

Scan Description SAGITTAL
3D PC Venogram
Pre
CE MRV
CORONAL
TWIST
Post
CE MRV
         
TR 61.7 ms 3.95 ms 2.38 ms 3.95 ms
TE 10.20 ms 1.49 ms 1.04 ms 1.49 ms
Flip Angle 15 deg 25 deg 15 deg 25 deg
Bandwidth 300 Hz/Px 450 Hz/Px 720 Hz/Px 450 Hz/Px
Fat-Water Contrast ― Standard Standard Standard
Gradient Mode Fast Fast Fast Fast
RF Pulse Type Normal Normal Fast Normal
Asymmetric Echo Weak Allowed Strong Allowed

Scan Parameters

Scan Description SAGITTAL
3D PC Venogram
Pre
CE MRV
TWIST
CORONAL
Post
CE MRV
         
Base Resolution 256 288 256 288
Phase Resolution 75% 82% 75% 82%
Slice Resolution 50% 70% 75% 70%
Averages 1 1 ― 1
Concatenations 1 ― ― ―
FoV Read 220 mm 240 mm 300 mm 240 mm
FoV Phase 100% 75% 56% 75%
Phase Encoding Dir. A >> P R >> L R >> L R >> L
Slice Oversampling 0% 20% 17% 20%
Phase Oversampling 0% 0% 0% 0%
Slice Partial Fourier ― Off 7/8 Off
Phase Partial Fourier Off 7/8 7/8 7/8
Flow Compensation None ― ― ―

Scanning Range

Scan Description SAGITTAL
3D PC Venogram
Pre
CE MRV
TWIST
CORONAL
Post
CE MRV
         
Elliptical Filter On Off On Off
Normalize Off Prescan Off Prescan
Distortion Correction 2D 2D 2D 2D
Distance Factor 20% 20% ― 20%
Slice Thickness 1.0 mm 1.0 mm 1.7 mm 1.0 mm
Slices per Slab 144 160 96 160
Time to center ― 6.0 s 6.1 s 6.0 s
Acquisition Time 8:01 min 27 sec 1:05 min 27 sec