Field Strength

MRA Head Without Contrast 3T

Siemens Healthineers – Non-Contrast MR Protocols

Get Adobe Reader

Instructions And Patient Preparation

Instructions
This sequence is in addition to the Routine Adult Brain protocol
Position
Supine, head first
Coil
20 channel Head/Neck coil or better
Landmark
Glabella

Imaging Parameters

Scan Description AXIAL
3D TOF
  View Image
Sequence Name fl_r
Sequence Type Gre
Dimension 3D
Acceleration Mode CS
AutoAlign Head > Brain

Scan Parameters

Scan Description AXIAL
3D TOF
   
TR 21.0 ms
TE 3.69 ms
Flip Angle 20 deg
Bandwidth 187 Hz/Px
Fat-Water Contrast Standard
Gradient Mode Fast
RF Pulse Type Normal
Asymmetric Echo Weak

Acquisition Parameters

Scan Description AXIAL
3D TOF
   
Base Resolution 352
Phase Resolution 95%
Slice Resolution 67%
Averages 1
Concatenations 6
FoV Read 220 mm
FoV Phase 91%
Phase Encoding Dir. R >> L
Phase Oversampling 0%
Slice Oversampling 25%
Slice Partial Fourier Off
Phase Partial Fourier Off
Flow compensation Slice/Read

Scanning Range

Scan Description AXIAL
3D TOF
   
Elliptical Filter Off
Filter Distortion Corr.(2D),
Prescan Normalize
Prescan Normalize On
Distance Factor -21%
Slice Thickness 0.4 mm
Slabs 6
Slices per Slab 48
Acquisition Time 4:31 min