Filters
Vendor
Field Strength

MRA Neck 3T

Siemens Healthineers – MR Gadopiclenol (VUEWAY®) Protocols

Get Adobe Reader

Instructions And Patient Preparation

Instructions
The following sequences are in addition to the Routine Adult Brain protocol
Position
Supine, head first
Coil
20 channel Head/Neck coil or better
Landmark
Glabella

Contrast Agent Parameters

Contrast Agent Dose Injection Rate
Gadopiclenol (VUEWAY) 0.05 mmol/kg CORONAL TWIST Inject 1-2 mL contrast
Inject remaining contrast CE MRV Post – 2 measurements
Saline Flush 20 mL 1-1.5 mL/sec (CEMRA) / 3-4 mL/sec (DSC Perfusion)
Comments • Report agent administered and dose
• Use minimal dose for patients with a GFR <60 mL/sec

Imaging Parameters

Scan Description AXIAL
3D TOF
AXIAL
T1 Fat Sat
Black Blood
(When requested)
Pre
CE MRA
CORONAL
TWIST
Inject 1-2 mL Contrast
Post
CE MRA
Inject Remaining Contrast
2 Measurements
  View Image View Image View Image View Image View Image
Sequence Name fl_r tse fl fldyn fl
Mode 2D 2D 2D 2D 2D
PAT Mode GRAPPA GRAPPA GRAPPA GRAPPA GRAPPA
Dimension 3D 2D 3D 3D 3D
Acceleration Factor PE 2 2 3 2 3
Acceleration Factor 3D 1 ― 1 2 1

Scan Parameters

Scan Description AXIAL
3D TOF
AXIAL
T1 Fat Sat
Black Blood
Pre
CE MRA
TWIST
TIMING
Post
CE MRA
           
TR 20.0 ms 746.0 ms 3.34 ms 2.19 ms 3.34 ms
TE 3.16 ms 13 ms 1.23 ms 0.89 ms 1.23 ms
Flip Angle 20 deg 180 deg 25 deg 18 deg 25 deg
Turbo Factor ― 3 ― ― ―
Bandwidth 250 Hz/Px 130 Hz/Px 590 Hz/Px 720 Hz/Px 590 Hz/Px
Fat Saturation Mode ― Strong ― ― ―
Fat Suppression None Fat sat. None None None
Water Suppression None None ― ― ―
Gradient Mode Fast Fast Fast Fast Fast
RF Pulse Type ― Normal Normal Fast Normal
Asymmetric Echo Allowed ― Allowed Strong Allowed

Acquisition Parameters

Scan Description AXIAL
3D TOF
AXIAL
T1 Fat Sat
Black Blood
Pre
CE MRA
TWIST
TIMING
Post
CE MRA
           
Base Resolution 384 448 448 288 448
Phase Resolution 65% 50% 82% 75% 82%
Slice Resolution 63% 100% 61% 75% 61%
Averages 1 ― 1 ― 1
Averaging Mode Short term Short term Short term ― Short term
Concatenations 8 3 ― ― ―
FoV Read 190 mm 250 mm 350 mm 350 mm 350 mm
FoV Phase 100% 78% 81% 75% 81%
Phase Encoding Dir. A >> P A >> P R >> L R >> L R >> L
Phase Oversampling 10% 75% 0% 0% 0%
Slice Oversampling 23% ― 13% 13% 13%
Phase Partial Fourier Off Off 6/8 6/8 6/8
Slice Partial Fourier 6/8 ― 6/8 6/8 6/8
Flow Compensation Yes No ― ― ―

Scanning Range

Scan Description AXIAL
3D TOF
AXIAL
T1 Fat Sat
Black Blood
Pre
CE MRA
TWIST
TIMING
Post
CE MRA
           
Elliptical Filter On Off Off On Off
Filter Distortion Corr.(2D),
Prescan Normalize,
Elliptical filter
Distortion Corr.(2D) Distortion Corr.(2D),
Prescan Normalize
Distortion Corr.(2D),
Elliptical filter
Distortion Corr.(2D),
Prescan Normalize
Normalize Off Off Off Off Off
Prescan Normalize On Off On Off On
Distortion Correction On On On On On
Distance Factor -35% 20% 20% ― 20%
Slice Thickness 1.0 mm 5.0 mm 0.75 mm 1.33 mm 0.75 mm
Slices 26 Slices per Slab 35 Slices 128 Slices per Slab 128 Slices per Slab 128 Slices per Slab
Acquisition Time 6:47 min 2:07 min 23 sec 1:09 min 46 sec