Instructions And Patient Preparation

Instructions
• Ensure to increase the FOV as needed to include all breast tissue and axillae; use same FOV for all sequences
• Ensure to pull abdominal tissue down to prevent a fold into the coil and FOV
• Ensure to pull medial breast tissue into coil
• Ensure axilla tissue is in the coil and not above the side panel
Common Indications
Silicone screening
Position
Prone
Coil
Bilateral Breast Coil
Landmark
Mid-breast, Nipples

Contrast Agent Parameters

Contrast Agent Dose Injection Rate
Gadopiclenol (VUEWAY) 0.05 mmol/kg up to a calculated volume of 10 mL (If adding Routine Bilateral) 2 mL/sec
Saline Flush 20 mL 2 mL/sec

Info Page

Scan Description SURVEY AXIAL
SILICONE
AXIAL
T2 F_S
LT
SILICONE
RT
SILICONE
           
Total scan duration 00:23.3 04:14.4 05:32.5 04:14.4 04:14.4
Rel. SNR 2.4 1.1 3 1.1 1.1
Act. TR (ms) — — 4750 — —
Act. TE (ms) — 60 74 60 60
Act. TR/TE (ms) 7.1 / 3.5 — — — —
Act. TR/TI (ms) — 10600 / 2700 — 10600 / 2700 10600 / 2700
ACQ matrix M x P 256 x 128 184 x 197 184 x 224 176 x 142 176 x 142
ACQ voxel MPS (mm) 1.76 / 3.52 / 10.0 1.25 / 1.56 / 3.00 1.25 / 1.37 / 3.00 1.25 / 1.55 / 3.00 1.25 / 1.55 / 3.00
REC voxel MPS (mm) 1.76 / 1.76 / 10.0 0.87 / 0.87 / 3.00 0.48 / 0.48 / 3.00 0.76 / 0.76 / 3.00 0.76 / 0.76 / 3.00
Scan percentage (%) 50 80 91 80 80
Packages — 2 2 2 2
Min. slice gap (mm) — 3 3 3 3
Act. slice gap (mm) — 0.3 — 0.3 0.3
Act. WFS (pix) / BW (Hz) 0.854 / 254.3 — — — —
Min. WFS (pix) / Max. BW (Hz) 0.280 / 775.0 — — — —
WFS (pix) / BW (Hz) — 0.796 / 272.8 0.336 / 647.0 0.794 / 273.7 0.794 / 273.7
TSE es / shot (ms) — 6.0 / 114 7.0 / 141 6.0 / 114 6.0 / 114
TEeff / TEequiv (ms) — — 74 / 53 — —
Min. TR (ms) — — 4750 — —
Min. TR/TE (ms) 7.1 / 3.1 — — — —
Local torso SAR < 21 % < 42 % < 50 % < 42 % < 42 %
Whole body SAR / level < 0.7 W/kg / normal < 1.4 W/kg / normal < 1.7 W/kg / normal < 1.4 W/kg / normal < 1.4 W/kg / normal
SED 0.0 kJ/kg < 0.4 kJ/kg < 0.6 kJ/kg < 0.4 kJ/kg < 0.4 kJ/kg
Max B1+rms 1.95 uT 2.75 uT 3.01 uT 2.75 uT 2.75 uT
PNS / level 24 % / normal 50 % / normal 60 % / normal 60 % / normal 60 % / normal
dB/dt 12.8 T/s 52.4 T/s 90.6 T/s 47.8 T/s 47.8 T/s
Sound Pressure Level (dB) 9.5 14.2 18.6 11.1 11.1

Geometry

Scan Description SURVEY AXIAL
SILICONE
AXIAL
T2 F_S
LT
SILICONE
RT
SILICONE
           
SmartSelect yes yes yes yes yes
Coil 1 (exclude) — None None None None
Uniformity — CLEAR CLEAR CLEAR CLEAR
FOV AP (mm) 450 230 230 220 220
FOV (mm) 290, (19) 82 — — — —
FOV FH (mm) 450 152 192 220 220
FOV RL (mm) — 307 307 152 152
ACQ voxel size AP (mm) 3.5 1.3 1.25 1.5 1.5
ACQ voxel size FH (mm) 1.8 — — 1.3 1.3
ACQ voxel size RL (mm) — 1.5 1.2 — —
Slice thickness (mm) 10 3 3 3 3
Bar orientation RL — — — —
VOI orientation transverse — — — —
VOI size AP (mm) 30 — — — —
VOI size RL (mm) 30 — — — —
VOI size FH (mm) 30 — — — —
Recon voxel size AP (mm) 1.8 0.9 0.5 0.8 0.8
Recon voxel size FH (mm) 1.8 — — 0.8 0.8
Recon voxel size RL (mm) — 0.9 0.5 — —
Image shutter yes — — — —
Fold-over suppression no oversampling oversampling oversampling oversampling
    L (mm) — 101.5 21.7 — —
    R (mm) — 101.5 21.7 — —
    P (mm) 50 — — 144.8 144.8
    A (mm) 50 — — 144.8 144.8
Reconstruction matrix 256 352 640 288 288
SENSE no no yes no no
    P reduction (RL) — — 1.6 — —
CS-SENSE no yes no yes yes
    reduction — 3.5 — 3.5 3.5
    denoising — system default — system default system default
Stacks 3 1 1 1 1
    type (20) parallel parallel parallel parallel parallel
Slices 15, (19) 5 46 64 46 46
    slice gap user defined default user defined default default
    gap (mm) 10, (19) 8 — 0 — —
Slice orientation sagittal, transverse, (18) coronal transverse transverse sagittal sagittal
Fold-over direction AP, (19) RL RL RL AP AP
Fat shift direction F, P, (18) F P P F F
    radial axis (2) AP, (18) RL — — — —
    radial angle (deg) (20) 0 — — — —
Minimum number of packages — 1 1 1 1
Slice scan order default default default default default
Stack scan order ascend — — — —
O-MAR no no no no no
Large table movement no no no no no
Stack alignment no — — — —
Stack sort order no — — — —
PlanAlign no no no no no
REST slabs 0 0 0 0 0
Interactive positioning no no no no no
Patient position feet first feet first feet first feet first feet first
Patient body position feet first feet first feet first feet first feet first
Patient orientation prone prone prone prone prone
Patient body orientation prone prone prone prone prone

Contrast

Scan Description SURVEY AXIAL SILICONE AXIAL T2 F_S LT SILICONE RT SILICONE
Scan type Imaging Imaging Imaging Imaging Imaging
Scan mode M2D MS MS MS MS
technique FFE IR SE IR IR
Modified SE — — no — —
Contrast enhancement T1 — — — —
Acquisition mode cartesian cartesian cartesian cartesian cartesian
Fast Imaging mode none TSE TSE TSE TSE
shot mode — multishot multishot multishot multishot
TSE factor — 19 20 19 19
startup Echoes — — — — —
profile order — linear linear linear linear
DRIVE — no no no no
ultrashort — no yes no no
fid reduction — default default default default
Echoes 1 1 1 1 1
partial echo no no no no no
shifted echo no — — — —
TE user defined user defined user defined user defined user defined
(ms) 3.5 60 74 60 60
Flip angle (deg) 50 — 90 — —
Refocusing control — no PSS Adv (T2-opt) no no
min. angle (deg) — — 70 — —
mid. angle (deg) — — 100 — —
max. angle (deg) — — 140 — —
TR shortest shortest range shortest shortest
minimum (ms) — — 2500 — —
maximum (ms) — — 5000 — —
Halfscan no no no no no
Water-fat shift maximum maximum maximum maximum maximum
IR delay (ms) — 2700 — 2700 2700
acquire during delay — yes — yes yes
dual — yes — yes yes
delay (ms) — 180 — 180 180
Shim — default default default default
mDIXON no no yes no no
Fat suppression no no no no no
Water suppression no no no no no
BB pulse — no no no no
MTC no no no no no
Diffusion mode no no no no no
T1 mapping no no no no no
SAR mode high high high high high
B1 mode default default default default default
SAR allow first level yes yes yes yes yes
Patient pregnancy no no no no no
Patient WB SAR [W/kg] — — — — —
Patient Head SAR [W/kg] — — — — —
Patient max. dB/dt [T/s] — — — — —
Max slewrate [T/m/s] — — — — —
Max. B1+rms [uT] — — — — —
PNS mode low low low low low
Gradient mode default default maximum default default

Motion

Scan Description SURVEY AXIAL
SILICONE
AXIAL
T2 F_S
LT
SILICONE
RT
SILICONE
           
Cardiac synchronization no no no no no
Respiratory compensation no no no no no
Navigator respiratory comp no no no no no
Flow compensation no no no no no
fMRI echo stabilisation no — — — —
Temporal slice spacing — — default — —
Motion smoothing — no yes no no
NSA 1 2 2 2 2
SMART — no yes no no
MRE enable no no no no no

Dynamic Angio

Scan Description SURVEY AXIAL
SILICONE
AXIAL
T2 F_S
LT
SILICONE
RT
SILICONE
           
Angio / Contrast enh. no — — — —
Quantitative flow no — — — —
Manual start no no no no no
Dynamic study no no no no no
Arterial Spin labeling no no no no no

Post Processing

Scan Description SURVEY AXIAL
SILICONE
AXIAL
T2 F_S
LT
SILICONE
RT
SILICONE
           
Preparation phases — full auto full full
Interactive F0 no no no no no
Quick Survey default default default default default
MIP/MPR no no no no no
SWIp no — — — —
mDIXON images — — water, (5) no — —
Images M, (3) no M, R, (2) no — M, R, (2) no M, R, (2) no
Autoview image M M — M M
Calculated images (4) no — — — —
Reference tissue Breast Breast Breast Breast Breast
Recon compression No No No No No
Preset window contrast soft soft soft soft soft
Reconstruction mode real time real time real time real time real time
Save raw data no no yes no no
Hardcopy protocol no no no no no
Image filter system default system default medium system default system default
Uniformity correction yes yes yes yes yes
Geometry correction default default default default default
Free rotatable no no no no no
IF_info_seperator 1634755923 1634755923 1634755923 1634755923 1634755923