Philips Healthcare – MR Gadopiclenol (VUEWAY®) Protocols

Get Adobe Reader

Instructions And Patient Preparation

Instructions
• Ensure to increase the FOV as needed to include all breast tissue and axillae; use same FOV for all sequences
• Ensure to pull abdominal tissue down to prevent a fold into the coil and FOV
• Ensure to pull medial breast tissue into coil
• Ensure axilla tissue is in the coil and not above the side panel
Common Indications
Screening
Position
Prone
Coil
Bilateral Breast Coil
Landmark
Mid-breast, Nipples

Contrast Agent Parameters

Contrast Agent Dose Injection Rate
Gadopiclenol (VUEWAY) 0.05 mmol/kg up to a calculated volume of 10 mL 2 mL/sec
Saline Flush 20 mL 2 mL/sec

Info Page

Scan Description SURVEY AXIAL
T1
AXIAL
T2 SPAIR
DYNAMICS
THRIVE
LT
OBLIQUE
SAGITTAL
RT
OBLIQUE
SAGITTAL
             
Total scan duration 00:32.0 01:42.7 04:14.6 03:30.5 02:04.1 01:59.1
Rel. SNR 6.3 2.7 8.6 1 4.3 4.1
Act. TR (ms) — — 4715 — — —
Act. TE (ms) — — 70 — — —
Act. TR/TE (ms) 9.5 / 4.6 5.2 / 2.7 — 3.7 / 1.90 4.2 / 2.2 4.2 / 2.2
Dyn. scan time — — — 01:09.2 — —
ACQ matrix M x P 232 x 232 268 x 402 144 x 219 160 x 243 220 x 220 220 x 220
ACQ voxel MPS (mm) 1.51 / 1.51 / 7.00 0.60 / 0.60 / 2.00 1.11 / 1.11 / 2.00 1.00 / 1.00 / 2.00 1.00 / 1.00 / 2.00 1.00 / 1.00 / 2.00
REC voxel MPS (mm) 1.46 / 1.46 / 7.00 0.25 / 0.25 / 2.00 0.56 / 0.56 / 2.00 0.34 / 0.34 / 2.00 0.51 / 0.51 / 2.00 0.57 / 0.57 / 2.00
Scan percentage (%) 100 99 100 100 100 100
Act. slice gap (mm) — 0 — 0 0 0
Packages — — 6 — — —
Min. slice gap (mm) — — 2 — — —
Optimal slices — — 30 — — —
Max.slices — — 90 — — —
TFE shots 2 630 — 336 536 526
TFE dur. shot / acq (ms) 426.5 / 369.3 160.7 / 160.7 — 206.0 / 124.8 206.4 / 114.1 196.4 / 104.3
TFE shot interval (ms) 426.5 160.7 — 206.0 228.6 223.5
Min. TI delay 229.3 — — — — —
Act. WFS (pix) / BW (Hz) 2.169 / 200.3 0.820 / 530.0 — 0.489 / 887.8 0.489 / 887.8 0.489 / 887.8
Min. WFS (pix) / Max. BW (Hz) 0.677 / 641.4 0.820 / 530.0 — 0.489 / 887.8 0.489 / 887.8 0.489 / 887.8
WFS (pix) / BW (Hz) — — 2.079 / 208.9 — — —
TSE es / shot (ms) — — 8.0 / 160 — — —
TEeff / TEequiv (ms) — — 70 / 52 — — —
Min. TR (ms) — — 4715 — — —
Min. TR/TE (ms) 9.5 / 3.9 — — — — —
SPAIR offset act./default (Hz) — — 220 [220] 220 [220] 220 [220] 220 [220]
Local torso SAR < 42 % < 24 % < 100 % < 60 % < 50 % < 50 %
Whole body SAR / level < 0.9 W/kg / normal < 0.5 W/kg / normal < 2.1 W/kg / 1st level < 1.3 W/kg / normal < 1.1 W/kg / normal < 1.1 W/kg / normal
SED 0.0 kJ/kg < 0.1 kJ/kg < 0.5 kJ/kg < 0.3 kJ/kg < 0.1 kJ/kg < 0.1 kJ/kg
Coil Power 40% 23% 96% 57% 48% 48%
Max B1+rms 1.47 uT 1.13 uT 2.30 uT 1.77 uT 1.62 uT 1.62 uT
PNS / level 19 % / normal 80 % / normal 47 % / normal 77 % / normal 59 % / normal 60 % / normal
dB/dt 16.1 T/s 58.4 T/s 53.8 T/s 60.7 T/s 32.4 T/s 32.4 T/s
Sound Pressure Level (dB) -4.58 16.87 7.44 9.39 11.44 9.67

Geometry

Scan Description SURVEY AXIAL
T1
AXIAL
T2 SPAIR
DYNAMICS
THRIVE
LT
OBLIQUE
SAGITTAL
RT
OBLIQUE
SAGITTAL
             
SmartSelect yes yes yes yes yes yes
Coil 1 (exclude) — None None None None None
Uniformity CLEAR CLEAR CLEAR CLEAR CLEAR CLEAR
FOV AP (mm) 350 160 160 160 220 220
FOV (mm) 245, (19) 75 — — — — —
FOV FH (mm) 350 172 172 172 220 220
FOV RL (mm) — 242 243 244 130 120
ACQ voxel size AP (mm) 1.5 0.6 1.1 1 1 1
ACQ voxel size FH (mm) 1.5 2 — 2 1 1
ACQ voxel size RL (mm) — 0.6 1 1 2 2
Slice thickness (mm) 7 — 2 — — —
Bar orientation RL — — — — —
VOI orientation transverse — — — — —
VOI size AP (mm) 30 — — — — —
VOI size RL (mm) 30 — — — — —
VOI size FH (mm) 30 — — — — —
Recon voxel size AP (mm) 1.5 0.3 0.6 0.3 0.5 0.6
Recon voxel size FH (mm) 1.5 2 — 2 0.5 0.6
Recon voxel size RL (mm) — 0.3 0.6 0.3 2 2
Fold-over suppression no oversampling oversampling oversampling oversampling oversampling
    L (mm) — 80.2 81.6 81.6 — —
    R (mm) — 80.2 81.6 81.6 — —
    F (mm) — — — — 100 100
    H (mm) — — — — 105 100
    A (mm) 50 — — — — —
    P (mm) 50 — — — — —
Slice oversampling — user defined — user defined default default
    oversample factor — 1 — 1 — —
ENCASE enable — no — no no no
Reconstruction matrix 240 960 432 720 432 384
SENSE no yes yes no yes yes
    P reduction (FH) — — — — 1 1
    P reduction (RL) — 2 2.3 — — —
    S reduction (FH) — 1 — — — —
    S reduction (RL) — — — — 1.7 1.7
CS-SENSE yes no no yes no no
    reduction 3 — — 1.9 — —
    denoising system default — — system default — —
k-t Acceleration no no no no no no
Stacks 3 1 1 1 1 1
    type (20) parallel — parallel — — —
Slices 15, (19) 5 86 86 86 65 60
    slice gap user defined — user defined — — —
    gap (mm) (20) 10 — 0 — — —
Slice orientation sagittal, transverse, (18) coronal transverse transverse transverse sagittal sagittal
Fold-over direction AP, (19) RL RL RL RL FH FH
Fat shift direction F, P, (18) F P P P P P
    radial axis (2) AP, (18) RL — — — — —
    radial angle (deg) (20) 0 — — — — —
Multi-chunk — no — no no no
Minimum number of packages — — 1 — — —
Slice scan order default — default — — —
Stack scan order ascend — — — — —
O-MAR — — no — — —
Move table per stack no — — — — —
Large table movement no no no no no no
Stack alignment no — — — — —
Stack sort order no — — — — —
PlanAlign no no no no no no
REST slabs 0 0 0 0 0 0
Interactive positioning no no no no no no
Patient position feet first feet first feet first feet first feet first feet first
Patient body position feet first feet first feet first feet first feet first feet first
Patient orientation prone prone prone prone prone prone
Patient body orientation prone prone prone prone prone prone

Contrast

Scan Description SURVEY AXIAL T1 AXIAL T2 SPAIR DYNAMICS THRIVE LT OBLIQUE SAGITTAL RT OBLIQUE SAGITTAL
Scan type Imaging Imaging Imaging Imaging Imaging Imaging
Scan mode M2D 3D MS 3D 3D 3D
technique FFE FFE SE FFE FFE FFE
Modified SE — — no — — —
Contrast enhancement T1 T1 — T1 T1 T1
Acquisition mode cartesian cartesian cartesian cartesian cartesian cartesian
Fast Imaging mode TFE TFE TSE TFE TFE TFE
shot mode multishot multishot multishot multishot multishot multishot
TFE factor 39 31 — 34 27 25
TSE factor — — 20 — — —
TE spacing — — user defined — — —
(ms) — — 8 — — —
startup Echoes default user defined — user defined user defined user defined
(number) — — — — — —
shot interval shortest shortest — shortest shortest shortest
profile order linear low_high asymmetric low_high low_high low_high
turbo direction — Z — radial Z Z
DRIVE — — no — — —
ultrashort — — no — — —
fid reduction — — default — — —
Echoes 1 1 1 1 1 1
partial echo no no no no no no
shifted echo no no — no no no
TE in-phase shortest user defined shortest shortest shortest
(ms) 4.6 — 70 — — —
Flip angle (deg) 25 12 90 12 12 12
Refocusing control — — T2-optimized — — —
min. angle (deg) — — 70 — — —
mid. angle (deg) — — 120 — — —
max. angle (deg) — — 160 — — —
TR shortest shortest range shortest shortest shortest
minimum (ms) — — 4000 — — —
maximum (ms) — — 5000 — — —
Halfscan no yes no yes yes yes
factor Y — 0.6 — 0.7 0.6 0.6
factor Z — 1 — 1 1 1
Water-fat shift maximum minimum maximum minimum minimum minimum
RF Shims fixed smart smart smart smart smart
Shim default SmartBreast SmartBreast SmartBreast SmartBreast SmartBreast
mDIXON no no no no no no
Fat suppression no no SPAIR SPAIR SPAIR SPAIR
power — — 2 2 2 2
inversion delay — — auto auto auto auto
suppr. level — — strong strong strong strong
frequency offset — — user defined user defined user defined user defined
offset (Hz) — — 220 220 220 220
Water suppression no no no no no no
TFE prepulse invert no — no no no
slice selection no — — — — —
shared no — — — — —
delay shortest — — — — —
PSIR no — — — — —
BB pulse — — no — — —
MTC no no no no no no
Diffusion mode no no no no no no
T1 mapping no no no no no no
SAR mode high high high high low low
B1 mode default default default default default default
SAR allow first level yes yes yes yes yes yes
Patient pregnancy no no no no no no
Patient WB SAR [W/kg] — — — — — —
Patient Head SAR [W/kg] — — — — — —
Patient max. dB/dt [T/s] — — — — — —
Max slewrate [T/m/s] — — — — — —
Max. B1+rms [uT] — — — — — —
PNS mode low moderate low moderate low low
Gradient mode regular default default default default default

Motion

Scan Description SURVEY AXIAL
T1
AXIAL
T2 SPAIR
DYNAMICS
THRIVE
LT
OBLIQUE
SAGITTAL
RT
OBLIQUE
SAGITTAL
             
Cardiac synchronization no no no no no no
Respiratory compensation no no no no no no
Navigator respiratory comp no no no no no no
Flow compensation no no no no no no
fMRI echo stabilisation no no — no no no
Temporal slice spacing — — default — — —
Motion smoothing — — no — — —
NSA 1 1 1 1 1 1
MRE enable no no no no no no

Dynamic Angio

Scan Description SURVEY AXIAL
T1
AXIAL
T2 SPAIR
DYNAMICS
THRIVE
LT
OBLIQUE
SAGITTAL
RT
OBLIQUE
SAGITTAL
             
Angio / Contrast enh. no no — no no no
Quantitative flow no no — no no no
CENTRA — no — no no no
Manual start no no no no no no
Dynamic study no no no individual no no
    dyn scans — — — 3 — —
    dyn scan times — — — user defined — —
    (mm:ss) — — — shortest (00:00.0),
manual (01:10.7),
shortest (02:21.3)
— —
    dummy scans — — — 0 — —
    immediate subtraction — — — no — —
    fast next scan — — — no — —
    synch. ext. device — — — no — —
    dyn stabilization — — — no — —
    prospect. motion corr. — — — no — —
Keyhole — — — no — —
Arterial Spin labeling no no no no no no

Post Processing

Scan Description SURVEY AXIAL
T1
AXIAL
T2 SPAIR
DYNAMICS
THRIVE
LT
OBLIQUE
SAGITTAL
RT
OBLIQUE
SAGITTAL
             
Preparation phases auto auto full auto full full
Interactive F0 no auto select auto select select
Quick Survey default default default default default default
MIP/MPR no no no no no no
SWIp no no — no no no
Images M, (3) no M, (3) no M, (3) no M, (3) no M, (3) no M, (3) no
Autoview image M M M M M M
Calculated images (4) no (4) no (4) no (4) no (4) no (4) no
Reference tissue Breast Breast Breast Breast Breast Breast
Recon compression No No No No No No
Preset window contrast soft soft soft soft soft soft
Reconstruction mode real time real time real time real time real time real time
Save raw data no no no no no no
Hardcopy protocol no no no no no no
Image filter system default system default system default system default system default system default
Uniformity correction no no yes no no no
Geometry correction default default default default default default
Elliptical k-space shutter — default — default default default
Free rotatable no no no no no no
IF_info_seperator 1634755923 1634755923 1634755923 1634755923 1634755923 1634755923