Brain Aneurysm 3T

Siemens Healthineers – MR Gadopiclenol (VUEWAY®) Protocols

Get Adobe Reader

Instructions And Patient Preparation

Instructions
• Exam is tailored to the patient’s condition and is meant for rapid assessment at lower resolution
• May stop scan whenever adequate clinical information is obtained
• Contrast only needed if MRA or dynamic susceptibility contrast (DSC) MR perfusion also obtained
Position
Supine, head first
Coil
20 channel Head/Neck coil or better
Landmark
Glabella

Contrast Agent Parameters

Contrast Agent Dose Injection Rate
Gadopiclenol (VUEWAY) 0.05 mmol/kg CORONAL TWIST Inject 1-2 mL contrast
Inject remaining contrast CE MRV Post – 2 measurements
Saline Flush 20 mL For DSC – Power Injector 3-4 mL/sec
Comments • For DSC – Inject after 5-6 phases
• Report agent administered and dose
• Use minimal dose for patients with a GFR <60 mL/sec

Imaging Parameters

Scan Description 3D TOF MRA Pre
CE MRA
Pre
HEAD SPACE
3D VWI
CORONAL
TWIST
Post
CE MRA
Post
HEAD SPACE
3D VWI
  View Image View Image View Image View Image View Image View Image
Sequence Name fl_r fl spcR fldyn fl spcR
Dimension 3D 3D 3D 3D 3D 3D
Acceleration Mode CS GRAPPA GRAPPA GRAPPA GRAPPA GRAPPA
Acceleration Factor PE ― 2 2 2 2 2
Acceleration Factor 3D ― 2 1 2 2 1
AutoAlign Head > Brain Head > Basis Head > Brain Head > Basis Head > Basis Head > Brain

Scan Parameters

Scan Description 3D TOF MRA Pre
CE MRA
Pre
HEAD SPACE
3D VWI
CORONAL
TWIST
Post
CE MRA
Post
HEAD SPACE
3D VWI
             
TR 21.0 ms 3.72 ms 700.0 ms 2.20 ms 3.72 ms 700.0 ms
TE 3.69 ms 1.33 ms 19.00 ms 0.91 ms 1.33 ms 19.00 ms
Flip Angle 20 deg 25 deg ― 15 deg 25 deg ―
Turbo Factor ― ― 38 ― ― 38
Bandwidth 187 Hz/Px 450 Hz/Px 630 Hz/Px 720 Hz/Px 450 Hz/Px 630 Hz/Px
Fat-Water Contrast Standard Standard Standard Standard Standard Standard
Gradient Mode Fast Fast Fast Fast Fast Fast
RF Pulse Type Normal Normal Normal Fast Normal Normal
Asymmetric Echo Weak Allowed ― Strong Allowed ―

Acquisition Parameters

Scan Description 3D TOF MRA Pre
CE MRA
Pre
HEAD SPACE
3D VWI
CORONAL
TWIST
Post
CE MRA
Post
HEAD SPACE
3D VWI
             
Base Resolution 352 288 256 256 288 256
Phase Resolution 95% 82% 90% 75% 82% 90%
Slice Resolution 67% 70% 90% 75% 70% 90%
Averages 1 1 1 ― 1 1
Concatenations 6 ― 1 ― ― 1
FoV Read 220 mm 240 mm 230 mm 300 mm 240 mm 230 mm
FoV Phase 91% 75% 100% 56% 75% 100%
Phase Encoding Dir. R >> L R >> L A >> P R >> L R >> L A >> P
Phase Oversampling 0 0% 10% % 0% 10%
Slice Oversampling 25% 20% 23% 17% 20% 23%
Slice Partial Fourier Off Off Off Off Off Off
Flow Compensation Slice/Read ― None ― ― None

Scanning Range

Scan Description 3D TOF MRA Pre
CE MRA
Pre
HEAD SPACE
3D VWI
CORONAL
TWIST
Post
CE MRA
Post
HEAD SPACE
3D VWI
             
Elliptical Filter Off Off Off On Off Off
Normalize Prescan Prescan Prescan Off Prescan Prescan
Distortion Correction 2D 2D 2D 2D 2D 2D
Slice Thickness 0.4 mm 1.0 mm 0.8 mm 1.7 mm 1.0 mm 0.8 mm
Distance Factor -21% 20% ― ― 20% ―
Slices 48 Slices per Slab 160 Slices per Slab 208 Slices per Slab 96 Slices per Slab 160 Slices per Slab 208 Slices per Slab
Acquisition Time 4:31 min 25 sec 7:58 min 1:05 min 25 sec 7:58 min