Brain MR Venography 3T

Philips Healthcare – MR Gadopiclenol (VUEWAY®) Protocols

Get Adobe Reader

Instructions And Patient Preparation

Instructions
• Sequences are in addition to the Routine Adult Brain protocol. For hyperacute patient, assess for penumbra
• No patient preparation
• 20- to 22-gauge IV access preferred
• Power injection preferred
Position
Supine, head first
Coil
SENSE-Head-8 channel coil preferred
Landmark
Glabella

Contrast Agent Parameters

Contrast Agent Dose Injection Rate Dose Timing Method and Scan Delay
Gadopiclenol (VUEWAY) 0.05 mmol/kg 1-1.5 mL/sec Hand bolus or power injection
Saline Flush 20 mL 1-1.5 mL/sec Hand bolus or power injection
Comments Report agent administered and dose

Info Page

Scan Description Survey AXIAL
TOF Venogram
Pre
CORONAL
TOF Venogram
Pre
AXIAL
TOF Venogram
+Gd
AXIAL
MPRAGE Venogram
+Gd
    View Image View Image View Image View Image
Total scan duration 00:31.4 02:06.2 02:06.2 02:06.2 04:13.8
Rel. SNR 1 0 0 0 0.4
Act. TR/TE(ms) 11 / 4.6 23 / 3.5 23 / 3.5 23 / 3.5 12 / 6.1
ACQ matrix M x P 268 x 128 260 x 260 260 x 260 260 x 260 420 x 420
ACQ voxel MPS (mm) 0.97 / 2.03 / 10.0 0.85 / 0.85 / 2.50 0.85 / 0.85 / 2.50 0.85 / 0.85 / 2.50 0.57 / 0.57 / 1.20
REC voxel MPS (mm) 0.90 / 0.90 / 10.0 0.42 / 0.42 / 2.50 0.42 / 0.42 / 2.50 0.42 / 0.42 / 2.50 0.56 / 0.56 / 1.20
Scan percentage (%) 48 100 100 100 100
Act. slice gap (mm) ― ― ― ― 0
TFE shots 2 ― ― ― 112
TFE dur. shot / acq(ms) 1162.5 / 704.0 ― ― ― 1756.2 / 1691.1
TFE shot interval (ms) 1162.5 ― ― ― ―
Min. TI delay 397.4 ― ― ― 897.511475
Act. WFS (pix) / BW (Hz) 3.101 / 140.1 3.198 / 135.8 3.198 / 135.8 3.198 / 135.8 3.969 / 109.4
Min. WFS (pix) / Max. BW (Hz) 0.587 / 740.3 0.542 / 801.3 0.542 / 801.3 0.542 / 801.3 0.919 / 472.4
Min. TR/TE(ms) 9.4 / 3.7 23 / 3.3 23 / 3.3 23 / 3.3 ―
Local torso SAR < 12 % < 62 % < 62 % < 62 % < 8 %
Whole body SAR / level < 0.4 W/kg / normal < 2.0 W/kg / normal < 2.0 W/kg / normal < 2.0 W/kg / normal < 0.3 W/kg / normal
SED 0.0 kJ/kg < 0.3 kJ/kg < 0.3 kJ/kg < 0.3 kJ/kg < 0.1 kJ/kg
Coil Power 0% 0% 0% 0% 0%
Max B1+rms 0.79 uT 1.80 uT 1.80 uT 1.80 uT 0.64 uT
PNS / level 53 % / normal 62 % / normal 90 % / 1st level 62 % / normal 66 % / normal
dB/dt 46.6 T/s 87.0 T/s 85.0 T/s 87.0 T/s 63.9 T/s
Sound Pressure Level (dB) 10.4 14.1 13.8 14.1 8.9
Boil-off (hPa/h) ― 3.6 5.4 3.6 3.6
Max Boil-off (hPa/h) ― 3.6 5.4 3.6 4.8
Boil-off (hPa) ― 0.1 0.2 0.1 0.3

Geometry

Scan Description Survey AXIAL
TOF Venogram
Pre
CORONAL
TOF Venogram
Pre
AXIAL
TOF Venogram
+Gd
AXIAL
MPRAGE Venogram
+Gd
           
Nucleus ― H1 H1 H1 H1
SmartSelect ― yes yes yes yes
Coil 1 (exclude) ― None None None None
Uniformity Classic CLEAR CLEAR CLEAR CLEAR
FOV AP (mm) 260 220 151 220 240
FOV FH (mm) 260 151 220 151 180
FOV RL (mm) ― 220 220 220 240
    stack RL (mm) 50 ― ― ― ―
ACQ voxel size AP (mm) 2 0.9 ― 0.9 0.6
ACQ voxel size FH (mm) 1 ― 0.9 ― 1.2
ACQ voxel size RL (mm) ― 0.8 0.8 0.8 0.6
Slice thickness (mm) 10 2.5 2.5 2.5 ―
Recon voxel size AP (mm) 0.9 0.4 ― 0.4 0.6
Recon voxel size FH (mm) 0.9 ― 0.4 ― 1.2
Recon voxel size RL (mm) ― 0.4 0.4 0.4 0.6
Fold-over suppression no no no no no
Slice oversampling ― ― ― ― user defined
    oversample factor ― ― ― ― 1.5
ENCASE enable ― ― ― ― no
Reconstruction matrix 288 528 528 528 432
SENSE no no no no no
CS-SENSE ― yes yes yes yes
    reduction ― 5 5 5 5
    denoising ― system default system default system default system default
k-t Acceleration ― no no no no
Stacks 3 1 1 1 1
    current A ― ― ― ―
    type parallel parallel parallel parallel ―
    slices 3 100 100 100 150
    slice gap user defined user defined user defined user defined ―
    gap (mm) 10 -1 -1 -1 ―
    slice orientation sagittal transverse coronal transverse transverse
    fold-over direction AP RL RL RL RL
    fat shift direction F P F P P
Slice scan order default FH PA FH ―
Stack scan order ascend ― ― ― ―
Stack sort order no ― ― ― ―
Multi-chunk ― ― ― ― no
Large table movement ― no no no no
Plan Align no no no no no
REST slabs 0 1 1 1 0
    type ― parallel perpendicular parallel ―
    thickness (mm) ― 20 20 20 ―
    position ― feet left feet ―
    gap ― user defined 100 user defined ―
    (mm) ― 10 90 10 ―
    power ― 1 1 1 ―
Interactive positioning ― no no no no
Allow table movement ― no no no no
Patient position ― head first head first head first head first
Patient body position ― head first head first head first head first
Patient orientation ― supine supine supine supine
Patient body orientation ― supine supine supine supine

Contrast

Scan Description Survey AXIALTOF VenogramPre CORONALTOF VenogramPre AXIALTOF Venogram+Gd AXIALMPRAGE Venogram +Gd
Scan type ― Imaging Imaging Imaging Imaging
Scan mode M2D M2D M2D M2D 3D
technique FFE FFE FFE FFE FFE
3D non-selective ― ― ― ― no
Contrast enhancement T1 T1 T1 T1 T1
Acquisition mode ― cartesian cartesian cartesian cartesian
Fast Imaging mode TFE none none none TFE
shot mode multishot ― ― ― multishot
TFE factor 64 ― ― ― 138
startup echoes default ― ― ― default
shot interval shortest ― ― ― user defined
(ms) ― ― ― ― 2250
profile order linear ― ― ― linear
turbo direction ― ― ― ― Z
Echoes 1 1 1 1 1
partial echo no yes yes yes no
shifted echo ― no no no no
TE in-phase out-phase out-phase out-phase shortest
(ms) 4.6 3.5 3.5 3.5 ―
Flip angle (deg) 15 60 60 60 9
TR user defined shortest shortest shortest shortest
(ms) 11 ― ― ― ―
Halfscan no no no no no
Water-fat shift user defined maximum maximum maximum maximum
(pixels) 3.5 ― ― ― ―
RF Shims fixed adaptive adaptive adaptive fixed
Shim default default default default default
mDIXON no no no no no
Fat suppression no no no no no
Water suppression no no no no no
TFE prepulse invert ― ― ― invert
slice selection no ― ― ― no
delay user defined ― ― ― user defined
(ms) 800 ― ― ― 900
PSIR ― ― ― ― no
MTC no no no no no
T2prep ― ― ― ― no
Custom prepulse ― no no no no
MDME ― no no no no
Diffusion mode ― no no no no
T1 mapping ― no no no no
Transmit channels ― both both both both
SAR mode high high high high high
B1 mode default user defined user defined user defined default
amplitude (uT) ― 11.5 11.5 11.5 ―
SAR allow first level yes no no no yes
Patient pregnancy ― yes yes yes no
Patient WB SAR [W/kg] ― — — — —
Patient Head SAR [W/kg] ― — — — —
Patient max. dB/dt [T/s] ― — — — —
Max slewrate [T/m/s] ― — — — —
Max. B1+rms [uT] ― — — — —
PNS mode low high high high high
Gradient mode default maximum maximum maximum default
SofTone mode no no no no no

Motion

Scan Description Survey AXIAL
TOF Venogram
Pre
CORONAL
TOF Venogram
Pre
AXIAL
TOF Venogram
+Gd
AXIAL
MPRAGE Venogram
+Gd
           
Cardiac synchronization no no no no no
Heart rate > 250 bpm ― no no no no
Respiratory compensation no no no no no
Navigator respiratory comp ― no no no no
Flow compensation no yes yes yes no
fMRI echo stabilisation ― no no no no
Motion smoothing ― ― ― ― no
NSA 1 1 1 1 1
MRE enable no no no no no

Dynamic Angio

Scan Description Survey AXIAL
TOF Venogram
Pre
CORONAL
TOF Venogram
Pre
AXIAL
TOF Venogram
+Gd
AXIAL
MPRAGE Venogram
+Gd
           
Angio / Contrast enh. no inflow inflow inflow no
Quantitative flow ― no no no no
CENTRA ― ― ― ― no
Manual start no no no no no
Dynamic study no no no no no
Arterial Spin labeling ― no no no no

Post Processing

Scan Description Survey AXIAL
TOF Venogram
Pre
CORONAL
TOF Venogram
Pre
AXIAL
TOF Venogram
+Gd
AXIAL
MPRAGE Venogram
+Gd
           
Preparation phases auto auto auto auto auto
Interactive F0 no no no no no
Quick Survey default default default default default
SmartPlan survey ― no no no no
B0 field map ― no no no no
B1 field map ― no no no no
MIP/MPR no MIP MIP MIP no
    protocol ― compose compose compose ―
SWIp ― no no no no
Images M, no, no, no M, (3) no M, (3) no M, (3) no M, (3) no
Autoview image M M M M M
Calculated images no, no, no, no (4) no (4) no (4) no (4) no
Reference tissue White matter Grey matter Grey matter Grey matter White matter
Recon compression No No No No No
Preset window contrast soft soft soft soft soft
Reconstruction mode immediate real time real time real time immediate
Save raw data no no no no no
Hardcopy protocol no no no no no
Image filter system default system default system default system default system default
Uniformity correction no no no no no
Geometry correction default default default default default
Elliptical k-space shutter ― ― ― ― default
Free rotatable no no no no no