Filters
Vendor
Field Strength

MR Venography (MRV) Abdomen Pelvis 1.5T

Siemens Healthineers – MR Gadopiclenol (VUEWAY®) Protocols

Get Adobe Reader

Instructions And Patient Preparation

Instructions
• Establish clear communication with the patient prior to the procedure regarding breath hold instructions
• Position arms close by the side to minimize the required field of view in the right-left direction
Common Indications
May-Thurner syndrome, vein occlusion, stenosis, blockage, vasculitis
Position
Supine, head first
Coil
Body Array Coil (two in tandem if available)
Landmark
Mid-thorax

Contrast Agent Parameters

Contrast Agents Dose Injection Rate Dose Timing Method and Scan Delay
Gadopiclenol (VUEWAY) 0.05 mmol/kg 1.5 mL/sec Fluoro trigger to peak arterial opacification of entire aorta
Saline Flush 20 mL 1.5 mL/sec Immediately following gadolinium
Comments • Report agent administered and dose
• Use minimal dose for patients with a GFR <30 mL/sec

Imaging Parameters

Scan Description AXIAL
T2 HASTE
AXIAL
T2 TRUFISP
Pre
AXIAL
VIBE
Post
AXIAL
VIBE
30 sec
Post
AXIAL
VIBE
60 sec
Post
AXIAL
VIBE
90 sec
Post
AXIAL
VIBE
180 sec
  View Image View Image View Image View Image View Image View Image View Image
Sequence Name h tfi fl fl fl fl fl
Dimension 2D 2D 3D 3D 3D 3D 3D
Acceleration mode GRAPPA GRAPPA CAIPIRINHA CAIPIRINHA CAIPIRINHA CAIPIRINHA CAIPIRINHA
CAIPIRINHA Mode ― ― Body Tra Body Tra Body Tra Body Tra Body Tra
Total Factor ― ― 4 4 4 4 4
Acceleration Factor PE 2 2 2 2 2 2 2
Acceleration Factor 3D ― ― 2 2 2 2 2

Scan Parameters

Scan Description AXIAL
T2 HASTE
AXIAL
T2 TRUFISP
Pre
AXIAL
VIBE
Post
AXIAL
VIBE
30 sec
Post
AXIAL
VIBE
60 sec
Post
AXIAL
VIBE
90 sec
Post
AXIAL
VIBE
180 sec
               
TR 1300.0 ms 3.8 ms 4.3 ms 4.3 ms 4.3 ms 4.3 ms 4.3 ms
TE 203.00 ms 1.92 ms 2.12 ms 2.12 ms 2.12 ms 2.12 ms 2.12 ms
Flip Angle 180 deg 90 deg 10 deg 10 deg 10 deg 10 deg 10 deg
Turbo Factor 198 Off Off Off Off Off Off
Bandwidth 698 Hz/Px 501 Hz/Px 390 Hz/Px 390 Hz/Px 390 Hz/Px 390 Hz/Px 390 Hz/Px
Fat-Water Contrast Standard Standard Fast Fat Saturation Fast Fat Saturation Fast Fat Saturation Fast Fat Saturation Fast Fat Saturation
Gradient Mode Fast Fast Fast Fast Fast Fast Fast
RF Pulse Type Fast Normal Fast Fast Fast Fast Fast
Asymmetric Echo ― Off Off Off Off Off Off

Acquisition Parameters

Scan Description AXIAL
T2 HASTE
AXIAL
T2 TRUFISP
Pre
AXIAL
VIBE
Post
AXIAL
VIBE
30 sec
Post
AXIAL
VIBE
60 sec
Post
AXIAL
VIBE
90 sec
Post
AXIAL
VIBE
180 sec
               
Base Resolution 256 256 256 256 256 256 256
Phase Resolution 95% 100% 75% 75% 75% 75% 75%
Slice Resolution ― ― 50% 50% 50% 50% 50%
Averages 1 2 1 1 1 1 1
Concatenations 3 ― 1 1 1 1 1
FoV Read 380 mm 380 mm 380 mm 380 mm 380 mm 380 mm 380 mm
FoV Phase 81% 81% 91% 91% 91% 91% 91%
Phase Encoding Dir. A >> P A >> P A >> P A >> P A >> P A >> P A >> P
Phase Oversampling 0% 0% 10% 10% 10% 10% 10%
Slice Oversampling ― ― 25% 25% 25% 25% 25%
Phase Partial Fourier Off Off Off Off Off Off Off
Slice Partial Fourier ― ― 6/8 6/8 6/8 6/8 6/8
Flow Compensation None None ― ― ― ― ―

Scanning Range

Scan Description AXIAL
T2 HASTE
AXIAL
T2 TRUFISP
Pre
AXIAL
VIBE
Post
AXIAL
VIBE
30 sec
Post
AXIAL
VIBE
60 sec
Post
AXIAL
VIBE
90 sec
Post
AXIAL
VIBE
180 sec
               
Raw Filter Off Off Off Off Off Off Off
Elliptical Filter Off Off Off Off Off Off Off
Distortion Correction 2D 2D 2D 2D 2D 2D 2D
Normalize Prescan Prescan Prescan Prescan Prescan Prescan Prescan
Image Filter Off Off Off Off Off Off Off
Distance Factor 20% 20% 20% 20% 20% 20% 20%
Slice Thickness 6.0 mm 4.0 mm 3.5 mm 3.5 mm 3.5 mm 3.5 mm 3.5 mm
Slices 40 Slices 30 Slices 96 Slices per Slab 96 Slices per Slab 96 Slices per Slab 96 Slices per Slab 96 Slices per Slab
Time of Acquisition 54 sec 42 sec 14 sec 14 sec 14 sec 14 sec 14 sec