Filters
Vendor
Field Strength

MRA Head Without Contrast 1.5T

Siemens Healthineers – Non-Contrast MR Protocols

Get Adobe Reader

Instructions And Patient Preparation

Instructions
This sequence is in addition to the Routine Adult Brain protocol
Position
Supine, head first
Coil
20 channel Head/Neck coil or better
Landmark
Glabella

Imaging Parameters

Scan Description 3D TOF MRA
  View Image
Sequence Name fl_r
Dimension 3D
Acceleration Mode GRAPPA
Acceleration Factor PE 2
Acceleration Factor 3D 1
AutoAlign Head > Brain

Acquisition Parameters

Scan Description 3D TOF MRA
   
TR 23.4 ms
TE 7.15 ms
Flip Angle 23 deg
Bandwidth 120 Hz/Px
Fat-Water Contrast Standard
Gradient Mode Fast
RF Pulse Type Normal
Asymmetric Echo Weak

Scan Parameters

Scan Description 3D TOF MRA
   
Base Resolution 320
Phase Resolution 80%
Slice Resolution 50%
Averages 1
Concatenations 6
FoV Read 200 mm
FoV Phase 100%
Phase Encoding Dir. R >> L
Slice Oversampling 33%
Phase Oversampling 20%
Slice Partial Fourier Off
Phase Partial Fourier Off
Flow Compensation Slice/Read

Scanning Range

Scan Description 3D TOF MRA
   
Raw Filter Off
Elliptical Filter Off
Distortion Correction 2D
Normalize B1 Filter
Image Filter Off
Distance Factor -31%
Slice Thickness 0.6 mm
Slabs 6
Slices per Slab 36
Acquisition Time 4:34 min