MRA Head 3T

Siemens Healthineers – MR Gadopiclenol (VUEWAY®) Protocols

Get Adobe Reader

Instructions And Patient Preparation

Instructions
• Sequences are in addition to Routine Adult Brain protocol
• For hyperacute patient, assess for penumbra
Position
Supine, head first
Coil
20 channel Head/Neck coil or better
Landmark
Glabella

Contrast Agent Parameters

Contrast Agent Dose Injection Rate
Gadopiclenol (VUEWAY) 0.05 mmol/kg CORONAL TWIST Inject 1-2 mL contrast
Inject remaining contrast CE MRV Post – 2 measurements
Saline Flush 20 mL 1-1.5 mL/sec
Comments • For bifurcations. 3D TOF if unable to do Carotid CEMRA or CEMRA is inadequate
• Report agent administered and dose
• Use minimal dose for patients with a GFR <60 mL/sec

Imaging Parameters

Scan Description 3D TOF MRA Pre
CE MRA
CORONAL
TWIST
Inject 1-2 mL Contrast
Post
CE MRA
Inject Remaining Contrast
2 Measurements
  View Image View Image View Image View Image
Sequence Name fl_r fl fldyn fl
Sequence Type Gre ― ― ―
Dimension 3D 3D 3D 3D
Acceleration Mode CS GRAPPA GRAPPA GRAPPA
Acceleration Factor PE ― 2 2 2
Acceleration Factor 3D ― 2 24 2
AutoAlign Head > Brain Head > Basis Head > Basis Head > Basis

Scan Parameters

Scan Description 3D TOF MRA Pre
CE MRA
CORONAL
TWIST
Post
CE MRA
         
TR 21.0 ms 3.72 ms 2.20 ms 3.72 ms
TE 3.69 ms 1.33 ms 0.91 ms 1.33 ms
Flip Angle 20 deg 25 deg 15 deg 25 deg
Bandwidth 187 Hz/Px 450 Hz/Px 720 Hz/Px 450 Hz/Px
Fat-Water Contrast Standard Standard Standard Standard
Gradient Mode Fast Fast Fast Fast
RF Pulse Type Normal Normal Fast Normal
Asymmetric Echo Weak Allowed Strong Allowed

Acquisition Parameters

Scan Description 3D TOF MRA Pre
CE MRA
CORONAL
TWIST
Post
CE MRA
         
Base Resolution 352 288 256 288
Phase Resolution 95% 82% 75% 82%
Slice Resolution 67% 70% 75% 70%
Averages 1 1 ― 1
Concatenations 6 ― ― ―
FoV Read 220 mm 240 mm 300 mm 240 mm
FoV Phase 91% 75% 56% 75%
Phase Encoding Dir. R >> L R >> L R >> L R >> L
Phase Oversampling 0% 0% 0% 0%
Slice Oversampling 25% 20% 17% 20%
Phase Partial Fourier Off 7/8 7/8 7/8
Slice Partial Fourier Off Off Off Off
Flow compensation Slice/Read ― ― ―

Scanning Range

Scan Description 3D TOF MRA Pre
CE MRA
CORONAL
TWIST
Post
CE MRA
         
Elliptical Filter Off Off On Off
Raw Filter Off Off Off Off
Distortion Correction 2D 2D 2D 2D
Distance Factor -21% 20% ― 20%
Slice Thickness 0.4 mm 1.0 mm 1.7 mm 1.0 mm
Slices per Slab 48 160 96 160
Acquisition Time 4:31 min 25 sec 1:05 min 25 sec