Field Strength

Multiple Sclerosis Brain 3T

Philips Healthcare – MR Gadopiclenol (VUEWAY®) Protocols

Get Adobe Reader

Instructions And Patient Preparation

Instructions
• Change patient into hospital issued gown and pants
• 18- to 22-gauge IV access
• Sagittal FLAIR using thin 3 mm slices
Position
Supine, head first
Coil
SENSE-Head-8 channel coil preferred. If a cervical spine MS evaluation is also done, then a 16 channel neurovascular coil is preferred
Landmark
Glabella

Contrast Agent Parameters

Contrast Agent Dose Injection Rate
Gadopiclenol (VUEWAY) 0.05 mmol/kg 1-2 mL/sec
Saline Flush 10-20 mL 1-2 mL/sec
Comments • 3D scan optional postcontrast – reformat orthogonal planes as appropriate
• Do 3D T2 FLAIR precontrast and reformat in orthogonal planes
• T2 postcontrast done to provide increased delay between contrast and Axial T1 postcontrast
• Postcontrast at least 1 plane should be similar sequence pre- and postcontrast

Info Page

Scan Description Survey
Brain
AXIAL
T1 +C
AXIAL
T2 Fat Sat
AXIAL
T2 FLAIR +C
Source
DTI
SAGITTAL
3D T1
SAGITTAL
3D T2 FLAIR
SAGITTAL
3D T1 +C
  View Image View Image View Image View Image View Image View Image View Image View Image
Total scan duration 00:18.9 02:16.8 01:45.0 02:45.0 02:00.8 02:17.8 03:44.0 02:17.8
Rel. SNR 1 1 1 1 1 1 1 1
Act. TR (ms) — 462 3000 — 4923 — — —
Act. TR/TI (ms) — — — 11000 / 2800 — — 8000 / 2400 —
Act. TE (ms) — 10 80 120 68 — 267 —
Act. TR/TE (ms) 15 / 3.2 — — — — 7.8 / 3.5 — 7.8 / 3.5
ACQ matrix M x P 124 x 84 276 x 180 224 x 175 316 x 175 112 x 112 220 x 220 240 x 240 220 x 220
ACQ voxel MPS (mm) 2.02 / 2.98 / 10.0 0.80 / 0.99 / 3.00 0.98 / 1.10 / 3.00 0.70 / 1.10 / 3.00 2.00 / 2.00 / 3.00 1.09 / 1.09 / 1.00 1.00 / 1.00 / 2.00 1.09 / 1.09 / 1.00
REC voxel MPS (mm) 1.74 / 1.74 / 10.0 0.43 / 0.43 / 3.00 0.46 / 0.46 / 3.00 0.43 / 0.43 / 3.00 1.00 / 1.00 / 3.00 0.94 / 0.94 / 1.00 0.50 / 0.50 / 1.00 0.94 / 0.94 / 1.00
Scan percentage (%) 68 80 89 63 100 100 100 100
Act. slice gap (mm) — — — — — 0 -1 0
Packages — 2 5 3 1 — — —
Min. slice gap (mm) — 0 0 0.6 0 — — —
Optimal slices — 25 10 34 — — — —
Max.slices — — — 51 — — — —
EPI factor — — — — 45 — — —
TFE shots 2 — — — — 68 — 68
TFE dur. shot / acq (ms) 701.6 / 630.0 — — — — 1752.0 / 1711.3 — 1752.0 / 1711.3
TFE shot interval (ms) 701.6 — — — — — — —
Min. TI delay 375.1 — — — — 890.3 — 890.3
Act. WFS (pix) / BW (Hz) 1.125 / 386.2 — — — — 2.080 / 208.9 — 2.080 / 208.9
Min. WFS (pix) / Max. BW (Hz) 0.509 / 854.3 — — — — 0.482 / 901.9 — 0.482 / 901.9
WFS (pix) / BW (Hz) — 2.676 / 162.3 1.619 / 268.3 1.999 / 217.3 13.193 / 32.9 — 0.475 / 913.7 —
BW in EPI freq. dir. (Hz) — — — — 1844.5 — — —
Full flow comp. — no — — — — no —
TSE es / shot (ms) — — 10.0 / 160 10.9 / 229 — — 4.3 / 503 —
TEeff / TEequiv (ms) — — 80 / 72 120 / 106 — — 267 / 128 —
Min. TR (ms) — 462 2865 — — — — —
Min. TR/TI (ms) — — — 7470 / 50 — — 2912 / 50 —
Min. TR/TE (ms) 6.3 / 3.2 — — — — — — —
SPAIR TR (ms) — — — — — — 8000 —
SPAIR inv. delay (ms) — — — — — — 220 —
SPAIR offset act./default (Hz) — — — — — — 100 [220] —
Local extremities SAR — < 72 % < 36 % < 21 % — < 3 % < 2 % < 3 %
Local torso SAR < 31 % — — — < 83 % — — —
Whole body SAR / level < 0.7 W/kg / normal < 0.2 W/kg / normal < 0.1 W/kg / normal < 0.1 W/kg / normal < 1.9 W/kg / normal 0.0 W/kg / normal 0.0 W/kg / normal 0.0 W/kg / normal
SED 0.0 kJ/kg 0.0 kJ/kg 0.0 kJ/kg 0.0 kJ/kg < 0.2 kJ/kg 0.0 kJ/kg 0.0 kJ/kg 0.0 kJ/kg
Coil Power — — — — 39% — — —
Max B1+rms 0.90 uT 4.23 uT 2.99 uT 2.38 uT 1.47 uT 0.87 uT 0.64 uT 0.87 uT
PNS / level 17 % / normal 74 % / normal 24 % / normal 20 % / normal 86 % / 1st level 51 % / normal 65 % / normal 51 % / normal
dB/dt 15.4 T/s 66.4 T/s 18.9 T/s 18.9 T/s 112.7 T/s 49.9 T/s 96.9 T/s 49.9 T/s
Sound Pressure Level (dB) 2.7 23 5 4.8 21.9 18.8 25.5 18.8

Geometry

Scan Description Survey
Brain
AXIAL
T1 +C
AXIAL
T2 Fat Sat
AXIAL
T2 FLAIR +C
Source
DTI
SAGITTAL
3D T1
SAGITTAL
3D T2 FLAIR
SAGITTAL
3D T1 +C
                 
SmartSelect yes yes yes yes yes yes yes yes
Coil 1 (exclude) — None None None None None None None
Uniformity — CLEAR CLEAR CLEAR CLEAR CLEAR CLEAR CLEAR
FOV AP (mm) 250 220 220 220 224 240 240 240
FOV (mm) (3) 50, (17) 10 — — — — — — —
FOV FH (mm) 250 150 150 150 150 240 240 240
FOV RL (mm) — 179 193 193 224 180 180 180
ACQ voxel size AP (mm) 2 0.8 1 0.7 2 1 1 1
ACQ voxel size FH (mm) 2 — — — — 1 1 1
ACQ voxel size RL (mm) — 1 1 0.9 2 1 2 1
Slice thickness (mm) 10 3 3 3 3 — — —
Bar orientation RL — — — — — — —
VOI orientation transverse — — — — — — —
VOI size AP (mm) 30 — — — — — — —
VOI size RL (mm) 30 — — — — — — —
VOI size FH (mm) 30 — — — — — — —
Recon voxel size AP (mm) 1.7 0.4 0.5 0.4 1 0.9 0.5 0.9
Recon voxel size FH (mm) 1.7 — — — — 0.9 0.5 0.9
Recon voxel size RL (mm) — 0.4 0.5 0.4 1 1 1 1
Fold-over suppression no no no no no no no no
      P (mm) 50 — — — — — — —
      A (mm) 50 — — — — — — —
Slice oversampling — — — — — default user defined default
      oversample factor — — — — — — 1 —
ENCASE enable — — — — — no — no
Reconstruction matrix 144 512 480 512 224 256 480 256
SENSE no no no no yes no no no
    P reduction (AP) — — — — 2.5 — — —
MB SENSE — — — — no — — —
CS-SENSE no yes yes yes no yes yes yes
      reduction — 2.5 2 2.2 — 3 6 3
      denoising — system default system default system default — system default system default system default
k-t Acceleration no no no no no no no no
Stacks 3 1 1 1 1 1 1 1
       type (20) parallel parallel parallel parallel parallel — — —
Slices (3) 3, (17) 1 50 50 50 50 180 180 180
      slice gap user defined user defined user defined user defined user defined — — —
      gap (mm) (20) 10 0 0 0 0 — — —
Slice orientation sagittal, coronal, (18) transverse transverse transverse transverse transverse sagittal sagittal sagittal
Fold-over direction AP, (19) RL RL RL RL AP AP AP AP
Fat shift direction (2) F, (18) P P P P P F F F
      radial axis AP, RL, AP, (17) RL — — — — — — —
      radial angle (deg) (20) 0 — — — — — — —
Minimum number of packages — 1 1 3 1 — — —
Slice scan order default default default interleaved default — — —
Stack scan order ascend — — — — — — —
Multi-chunk — — — — — no no no
O-MAR — no no no — — no —
Move table per stack no — — — — — — —
Large table movement no no no no no no no no
Stack alignment no — — — — — — —
Stack sort order no — — — — — — —
PlanAlign no no no no no no no no
REST slabs 0 0 0 0 0 0 0 0
      orientation — — — — transverse — — —
Patient position head first head first head first head first head first head first head first head first
Patient body position head first head first head first head first head first head first head first head first
Patient orientation supine supine supine supine supine supine supine supine
Patient body orientation supine supine supine supine supine supine supine supine

Contrast

Scan Description SurveyBrain AXIALT1 +C AXIALT2 Fat Sat AXIALT2 FLAIR +C SourceDTI SAGITTAL3D T1 SAGITTAL3D T2 FLAIR SAGITTAL3D T1 +C
Scan type Imaging Imaging Imaging Imaging Imaging Imaging Imaging Imaging Scan type —
Scan mode M2D MS MS MS MS 3D 3D 3D Scan mode —
Modified SE — no no — no — — — Modified SE —
3D non-selective — — — — — no — no 3D non-selective —
Contrast enhancement T1 — — — — T1 — T1 Contrast enhancement —
Acquisition mode cartesian cartesian cartesian cartesian cartesian cartesian cartesian cartesian Acquisition mode —
Fast Imaging mode TFE none TSE TSE EPI TFE TSE TFE Fast Imaging mode —
3D VIEW — — — — — — no — 3D VIEW —
TFE factor 42 — — — — 220 — 220 TFE factor —
TSE factor — — 15 21 — — 110 — TSE factor —
3D non-selective — — — — — no — no 3D non-selective —
Echoes 1 1 1 1 1 1 1 1 Echoes —
TE shortest user defined user defined user defined shortest shortest shortest shortest TE —
Flip angle (deg) 20 90 90 — 90 8 — 8 Flip angle (deg) —
Refocusing control — — constant constant — — constant — Refocusing control —
TR user defined range user defined user defined shortest shortest user defined shortest TR —
Halfscan no no no no yes no — no Halfscan —
Water-fat shift maximum maximum maximum user defined minimum maximum maximum maximum Water-fat shift —
IR delay (ms) — — — 2800 — — 2400 — IR delay (ms) —
RF Shims — — — — fixed — — — RF Shims —
Shim — default default default PB-volume auto default auto Shim —
ShimAlign — — — — no — — — ShimAlign —
mDIXON no no no no no no no no mDIXON —
Fat suppression no no SPAIR no SPIR no SPAIR no Fat suppression —
Grad Rev Fat suppr — — — — no — — — Grad Rev Fat suppr —
Water suppression no no no no no no no no Water suppression —
TFE prepulse invert — — — — invert — invert TFE prepulse —
BB pulse — no no no no — no — BB pulse —
MTC no no no no no no no no MTC —
Diffusion mode no no no no DTI no no no Diffusion mode —
T1 mapping no no no no no no no no T1 mapping —
SAR mode high high high high high high high high SAR mode —
B1 mode default default default default default default default default B1 mode —
SAR allow first level yes yes yes yes yes yes yes yes SAR allow first level —
PNS mode low high low low high moderate high moderate PNS mode —
Gradient mode regular maximum default default maximum default maximum default Gradient mode —
SofTone mode no no user defined user defined no no no no SofTone mode —

Motion

Scan Description Survey
Brain
AXIAL
T1 +C
AXIAL
T2 Fat Sat
AXIAL
T2 FLAIR +C
Source
DTI
SAGITTAL
3D T1
SAGITTAL
3D T2 FLAIR
SAGITTAL
3D T1 +C
                 
Cardiac synchronization no no no no no no no no
Respiratory compensation no no no no no no no no
Navigator respiratory comp no no no no no no no no
Flow compensation no yes no no no no sensitized no
fMRI echo stabilisation no — — — — no — no
Temporal slice spacing — default default — default — — —
Motion smoothing — — no no — — no —
NSA 1 2 1 1 1 1 1 1
SMART — no — — — — — —
MRE enable no no no no no no no no

Dynamic Angio

Scan Description Survey
Brain
AXIAL
T1 +C
AXIAL
T2 Fat Sat
AXIAL
T2 FLAIR +C
Source
DTI
SAGITTAL
3D T1
SAGITTAL
3D T2 FLAIR
SAGITTAL
3D T1 +C
                 
Angio / Contrast enh. no — — — — no — no
Quantitative flow no — — — — no — no
CENTRA — — — — — no no no
Manual start no no no no no no no no
Dynamic study no no no no no no no no
      dyn stabilization — — — — regular — — —
Arterial Spin labeling no no no no no no no no

Post Processing

Scan Description Survey
Brain
AXIAL
T1 +C
AXIAL
T2 Fat Sat
AXIAL
T2 FLAIR +C
Source
DTI
SAGITTAL
3D T1
SAGITTAL
3D T2 FLAIR
SAGITTAL
3D T1 +C
                 
Preparation phases — auto auto auto auto auto auto auto
Interactive F0 no no no no no no no no
Quick Survey default default default default default default default default
B0 field map no no no no no no no no
MIP/MPR no no no no no no no no
SWIp no — — — — no — no
Images M, (3) no M, (3) no M, (3) no M, (3) no M, (3) no M, (3) no M, (3) no M, (3) no
Autoview image M M M M M M M M
Calculated images (4) no (4) no (4) no — (4) no (4) no — (4) no
Reference tissue White matter Grey matter Grey matter White matter Grey matter White matter Grey matter White matter
Recon compression No No No No No No No No
Preset window contrast soft soft soft soft soft soft soft soft
Reconstruction mode immediate real time real time real time immediate immediate real time immediate
Save raw data no no no no no no no no
Hardcopy protocol no no no no no no no no
Image filter system default system default system default system default system default system default system default system default
Uniformity correction no no no no no no no no
Geometry correction default default default default default default default default
Elliptical k-space shutter — — — — — default default default
EPI geometry correction — — — — no — — —
IF_info_seperator — 1634755923 0 0 0 1634755923 1634755923 1634755923