Filters
Vendor
Field Strength

Orbits 1.5T

Siemens Healthineers – MR Gadopiclenol (VUEWAY®) Protocols

Get Adobe Reader

Instructions And Patient Preparation

Instructions
• These sequences usually done in addition to Routine Adult Brain protocol
• Have patient close eyes for these sequences
• Cover orbits through optic nerve
Position
Supine, head first
Coil
Head/Neck coil
Landmark
Glabella

Contrast Agent Parameters

Contrast Agent Dose Injection Rate
Gadopiclenol (VUEWAY) 0.05 mmol/kg Hand Bolus
Saline Flush 10 mL Hand Bolus
Comments • Consider using TSE DIXON to reduce susceptibility problems
• Report agent administered and dose
• Use minimal dose for patients with a GFR <60 mL/sec

Imaging Parameters

Scan Description CORONAL
STIR
CORONAL
T1
AXIAL
T2
AXIAL
BLADE DWI
Post
AXIAL
T1 Fat Sat
Post
CORONAL
T1 Fat Sat
Post
SAGITTAL
3D FLAIR SPACE
  View Image View Image View Image View Image View Image View Image View Image
Sequence Name tir tse tseR resolve tse tse spcir
Dimension 2D 2D 2D 2D 2D 2D 3D
Acceleration Mode GRAPPA GRAPPA GRAPPA GRAPPA GRAPPA GRAPPA CAIPIRINHA
Acceleration Factor PE 3 3 3 2 3 3 2
Acceleration Factor 3D ― ― ― ― ― ― 2
AutoAlign Head > Optic Nerves Head > Optic Nerves Head > Orbits Head > Orbits Head > Orbits Head > Optic Nerves Head > Basis

Acquisition Parameters

Scan Description CORONAL
STIR
CORONAL
T1
AXIAL
T2
AXIAL
BLADE DWI
Post
AXIAL
T1 Fat Sat
Post
CORONAL
T1 Fat Sat
Post
SAGITTAL
3D FLAIR SPACE
               
TR 3000.0 ms 670.0 ms 3760.0 ms 6000.0 ms 500.0 ms 630.0 ms 5000.0 ms
TE 49.00 ms 8.20 ms 97.00 ms TE 1: 59 ms
TE 2: 94 ms
8.70 ms 8.70 ms 338.00 ms
TI 160 ms ― ― ― ― ― 1600 ms
Flip Angle 150 deg 150 deg 150 deg 180 deg 150 deg 150 deg Mode T2 Var
Turbo Factor 10 2 17 ― 2 2 249
Bandwidth 200 Hz/Px 209 Hz/Px 171 Hz/Px ― 191 Hz/Px 191 Hz/Px None
EPI Factor ― ― ― 75 ― ― ―
Fat-Water Contrast Standard Standard Standard Fat Saturation Fat Saturation Fat Saturation Fat Saturation
Fat Saturation ― ― ― Strong Strong Strong Strong
Gradient Mode Normal Normal Normal Fast Normal Normal Fast
RF Pulse Type Normal Fast Fast Normal Fast Fast Non-sel.

Scan Parameters

Scan Description CORONAL
STIR
CORONAL
T1
AXIAL
T2
AXIAL
BLADE DWI
Post
AXIAL
T1 Fat Sat
Post
CORONAL
T1 Fat Sat
Post
SAGITTAL
3D FLAIR SPACE
               
Base Resolution 288 288 352 150 288 288 256
Phase Resolution 75% 86% 80% 100% 84% 80% 90%
Slice Resolution ― ― ― ― ― ― 70%
Averages 2 2 2 ― 2 2 2
Concatenations 2 1 1 1 2 2 1
FoV Read 170 mm 170 mm 170 mm 200 mm 170 mm 170 mm 260 mm
FoV Phase 100% 100% 100% 100% 100% 100% 88%
Phase Encoding Dir. R >> L R >> L R >> L A >> P R >> L R >> L A >> P
Slice Oversampling ― ― ― ― ― ― 9%
Phase Oversampling 60% 80% 100% 0% 80% 60% 0%
Slice Partial Fourier ― ― ― ― ― ― Off
Phase Partial Fourier Off Off Off ― Off Off Allowed
Flow Compensation None None None 667 Hz/Px None None Fast

Scanning Range

Scan Description CORONAL
STIR
CORONAL
T1
AXIAL
T2
AXIAL
BLADE DWI
Post
AXIAL
T1 Fat Sat
Post
CORONAL
T1 Fat Sat
Post
SAGITTAL
3D FLAIR SPACE
               
Raw Filter Off Off Off On Off Off On
Elliptical Filter Off Off Off ― Off Off Off
Distortion Correction 2D 2D 2D Off 2D 2D 2D
Normalize Image Based Prescan Prescan Prescan Prescan Prescan Prescan
Image Filter Off Off Off ― Off Off Off
Distance Factor 18% 18% 10% 0% 10% 18% ―
Slice Thickness 3.0 mm 3.0 mm 3.0 mm 3.0 mm 3.0 mm 3.0 mm 1.0 mm
Slices 28 Slices 28 Slices 21 Slices 21 Slices 21 Slices 28 Slices 176 Slices per Slab
Acquisition Time 2:56 min 1:46 min 1:39 min 3:20 min 2:32 min 2:44 min 5:07 min