Routine Adult Brain 3.0T

GE Healthcare – MR Gadopiclenol (VUEWAY®) Protocols

Get Adobe Reader

Instructions And Patient Preparation

Instructions
• Contrast is used for assessment of the blood-brain barrier, infection, tumor, and some stroke evaluation
• T2* GRE susceptibility sequence used for trauma, evaluation for blood, and seizure cases
• Perfusion sequence is optional in stroke or tumor cases
Position
Supine, head first
Coil
Head, HNU (Head 24 selected if possible)
Landmark
Glabella or chin

Contrast Agent Parameters

Contrast Agent Dose Injection Rate
Gadopiclenol (VUEWAY) 0.05 mmol/kg up to a calculated volume of 10 mL Hand Bolus usually
For perfusion-weighted imaging (PWI), Use 0.1 mmol/kg up to a calculated volume of 15 mL @ 4 mL/sec
Saline Flush 10-20 mL —
Comments • Report agent administered and dose
• Consider MPR 3 mm recons of all 3D sequences in all non-acquired planes

Imaging Parameters

Scan Description 3PL
LOCALIZER
SAGITTAL
T1 FLAIR
AXIAL
T2 FS FSE
AXIAL
T1 3D BRAVO
AXIAL
DWI
AXIAL
T2* GRE
Pre or Post
SAGITTAL
T2 FLAIR CUBE
Post
AXIAL
T1 3D BRAVO
Post
CORONAL
3D T1 CUBE
Optional:
SAGITTAL
3D T2 CUBE
Optional:
AXIAL
T2* SWAN
    View Image     View Image   View Image View Image View Image    
Primary Sequence Yes Yes Yes Yes Yes Yes Yes Yes Yes T1 Alternative T2 Alternative 3D T2*
Mode 2D 2D 2D 3D BRAVO 2D 2D 3D 3D BRAVO 3D 3D 3D
Pulse Sequence SSFSE IR FRFSE-XL BRAVO SE-EPI GRE CUBE BRAVO CUBE CUBE GRE
Imaging Options SEQ, EDR, FAST,
SS, ARC, TRF
FC, TRF,
SEQ, FAST
FC, TRF FAST, IRP,
ZIP2, ASSET
EPI, DIFF,
ASSET
FC ARC, T2-IRP,
ZIP2, EDR
FAST, IRP,
ZIP2, ASSET
EDR, ZIP2,
ZIP512, FR, ARC
EDR, FAST, FR, ZIP2, ARC FC, ASSET,
ZIP2, FAST

Scan Parameters

Scan Description 3PL
LOCALIZER
SAGITTAL
T1 FLAIR
AXIAL
T2 FS FSE
AXIAL
T1 3D BRAVO
AXIAL
DWI
AXIAL
T2* GRE
Pre or Post
SAGITTAL
T2 FLAIR CUBE
Post
AXIAL
T1 3D BRAVO
Post
CORONAL
3D T1 CUBE
Optional:
SAGITTAL
3D T2 CUBE
Optional:
AXIAL
T2* SWAN
                       
TE 80 24 80 ― Min 25 Fixed with TR ― 24 82 44
TR 1000 3500 3500-4250 ― 9000 800 6000 ― 550 2500 ―
TI ― Auto ― 450 ― ― 1954
(Fixed with TR)
450 ― ― ―
Flip Angle ― Auto Auto 13 ― 20 Auto 13 ― ― 15
ETL ― 8 19 ― ― ― 160 ― 24 100 ―
Bandwidth 83 41 41 25 125 31 32 25 62 62 62
Sat ― ― Fat ― ― ― Fat ― Fat ― ―

Acquisition Parameters

Scan Description 3PL
LOCALIZER
SAGITTAL
T1 FLAIR
AXIAL
T2 FS FSE
AXIAL
T1 3D BRAVO
AXIAL
DWI
AXIAL
T2* GRE
Pre or Post
SAGITTAL
T2 FLAIR CUBE
Post
AXIAL
T1 3D BRAVO
Post
CORONAL
3D T1 CUBE
Optional:
SAGITTAL
3D T2 CUBE
Optional:
AXIAL
T2* SWAN
                       
MATRIX freq 256 352 384 256 128 320 256 256 256 320 288
MATRIX phase 192 224 256 256 128 256 256 256 256 320 224
NEX 1 2 2 1 1 1 1 1 1 1 160
Phase FOV 1 1 0.75 ― 1 0.75 ― ― 0.7 1 1
Freq Direction Unswap S/I A/P A/P R/L A/P S/I A/P S/I S/I 1
Image Enhance ― PURE PURE PURE PURE PURE PURE PURE PURE PURE PURE
FC Direction ― ― Slice ― ― Slice ― ― ― ― ―
User CV ― CV12-1
CV22-1
― ACC=2 DIFF Dir-All
#T2 images
1 Nex 1 for T2
B1000/1 Nex
OPT TE On
Dual SE On
ACC=2
Ramp Samp 1
Recon 1
Grad OPT 1
― ACC=2
SL=2
ACC=2 CUBE Enh 3 ACC=2 ACC=2
OPT HS 1.2
Phase Images ACC=2 (HyperSense 1.2
if available)

Scanning Range

Scan Description 3PL
LOCALIZER
SAGITTAL
T1 FLAIR
AXIAL
T2 FS FSE
AXIAL
T1 3D BRAVO
AXIAL
DWI
AXIAL
T2* GRE
Pre or Post
SAGITTAL
T2 FLAIR CUBE
Post
AXIAL
T1 3D BRAVO
Post
CORONAL
3D T1 CUBE
Optional:
SAGITTAL
3D T2 CUBE
Optional:
AXIAL
T2* SWAN
                       
FOV 30 24 22 25 24 22 25 25 24 25 22
Slice Thickness 7 3 4 2.4 5 5 1.6 1.2 1.8 1.6 3
Spacing 5 1 1 ― 1 .5 ― ― ― ― ―
Number of Slices 7-Ax/Sag/Cor ― 24-26 70 24-26 24-26 126 140 160 112 24-26
Scan Time 0:18 sec 2-3 min ~2-3 min 2:30 min :50 sec 2 min 3-5 min 4 min 3-5 min 2:30 min 3-5 min