Filters
Vendor
Field Strength

Soft Tissue Neck 1.5T

Siemens Healthineers – MR Gadopiclenol (VUEWAY®) Protocols

Get Adobe Reader

Instructions And Patient Preparation

Instructions
• Flow Compensation is an option for all sequences
• Use pre- and postcontrast T1 when using Flow Compensation
• If contrast is used, lesions and cancer staging slices may vary based on indication
• Coronal coverage – top of orbits thru manubrium
• Axial coverage – tailored to area of interest (nasopharynx/oropharynx/parotids) cover lesions or area of interest
Position
Supine, head first
Coil
Head/Neck coil
Landmark
Chin

Contrast Agent Parameters

Contrast Agent Dose Injection Rate
Gadopiclenol (VUEWAY) 0.05 mmol/kg Hand Bolus
Saline Flush 10 mL Hand Bolus
Comments • No Fat Sat on precontrast
• Use Fast DIXON with Strong Fat Sat. Use Interpolation
• Report agent administered and dose
• Use minimal dose for patients with a GFR <60 mL/sec

Imaging Parameters

Scan Description SAGITTAL
T1
Pre
AXIAL
T1
AXIAL
T2 Fat Sat
DIXON
AXIAL
DWI
CORONAL
T2
Post
AXIAL
T1 DIXON
Post
CORONAL
T1 DIXON
Post
SAGITTAL
T1 DIXON
  View Image View Image View Image View Image View Image View Image View Image View Image
Sequence Name tseBR tse tseR epse tseBR tse tse tse
Dimension 2D 2D ― ― 2D ― ― ―
Acceleration Mode GRAPPA GRAPPA None GRAPPA GRAPPA GRAPPA GRAPPA GRAPPA
Acceleration Factor PE 2 2 ― 2 2 2 2 2
AutoAlign ― ― ― Head > Brain ― ― ― ―

Acquisition Parameters

Scan Description SAGITTAL
T1
Pre
AXIAL
T1
AXIAL
T2 Fat Sat
DIXON
AXIAL
DWI
CORONAL
T2
Post
AXIAL
T1 DIXON
Post
CORONAL
T1 DIXON
Post
SAGITTAL
T1 DIXON
                 
TR 3240.0 ms 595.0 ms 4070.0 ms 4900.0 ms 3460.0 ms 611.0 ms 611.0 ms 611.0 ms
TE 87.00 ms 17.00 ms 77.00 ms 98.0 ms 87.00 ms 13.00 ms 13.00 ms 13.00 ms
Flip Angle 160 deg 160 deg 180 deg ― 122 deg 180 deg 180 deg 180 deg
Turbo Factor 23 3 19 ― 23 3 3 3
Bandwidth 252 Hz/Px 200 Hz/Px 340 Hz/Px 1184 Hz/Px 252 Hz/Px 252 Hz/Px 252 Hz/Px 252 Hz/Px
Echo Spacing 7.28 ms 8.30 ms 11.02 ms ― 7.28 ms 13.46 ms 13.46 ms 13.46 ms
Echo Trains per Slice 32 68 22 ― 32 72 72 72
Fat-Water Contrast Standard Standard Dixon Fat Saturation Standard Dixon Dixon Dixon
Fat Saturation ― ― Strong Weak ― Strong Strong Strong
Gradient Mode Fast Fast Fast Fast Fast Fast Fast Fast
RF Pulse Type Fast Fast Fast Normal Fast Normal Normal Normal
Asymmetric Echo ― ― Off ― ― Off Off Off

Scan Parameters

Scan Description SAGITTAL
T1
Pre
AXIAL
T1
AXIAL
T2 Fat Sat
DIXON
AXIAL
DWI
CORONAL
T2
Post
AXIAL
T1 DIXON
Post
CORONAL
T1 DIXON
Post
SAGITTAL
T1 DIXON
                 
Base Resolution 320 320 320 128 320 320 320 320
Phase Resolution ― 70% 70% 100% ― 70% 70% 70%
Averages 1 3 1 ― 1 1 1 1
Concatenations 2 2 2 1 2 3 3 3
b-value ― ― ― 0 s/mm2
1000 s/mm2
― ― ― ―
FoV Read 200 mm 200 mm 200 mm 220 mm 220 mm 200 mm 220 mm 220 mm
FoV Phase 100% 100% 100% 100% 100% 100% 100% 100%
Phase Encoding Dir. A >> P R >> L R >> L A >> P R >> L R >> L R >> L A >> P
Phase Oversampling 100% 80% 80% 0% 100% 80% 80% 80%
Phase Partial Fourier ― Off Off ― ― Off Off Off
Flow Compensation None None None ― None None None None

Scanning Range

Scan Description SAGITTAL
T1
Pre
AXIAL
T1
AXIAL
T2 Fat Sat
DIXON
AXIAL
DWI
CORONAL
T2
Post
AXIAL
T1 DIXON
Post
CORONAL
T1 DIXON
Post
SAGITTAL
T1 DIXON
                 
Elliptical Filter Off On Off Off Off Off Off Off
Normalize Prescan Prescan Prescan Prescan Prescan Prescan Prescan Prescan
Distortion Correction 2D 2D 2D Off 2D 2D 2D 2D
Distance Factor 20% 30% 30% 0% 20% 30% 30% 20%
Slice Thickness 5.0 mm 5.0 mm 5.0 mm 5.0 mm 5.0 mm 5.0 mm 5.0 mm 5.0 mm
Slices 36 36 36 30 Slices 36 36 36 36
Acquisition Time 3:35 min 4:05 min 6:08 min 51 sec 3:50 min 4:27 min 4:27 min 4:27 min